    We got a 321GLtp a few months ago, but I'm just now getting a chance to play with its tactile capabilities. I think I have the tool calibrations right, but I'm not sure what to expect.
    I did a part alignment with a touch probe then measured the same circle and Z point with both the vision and the touch probe.
    The diameter measured exactly the same with both methods. For its center point, the X is the same, the Y is .0002 off and the Z is .0125 off.
    Is that an expected error on the XY axis?
    Can the Z ever be calibrated to be the same or are there too many variables with lighting to make that possible?
    Vision to Tactile (probe) Calibration chain should be as followed:
    Vision - Calibrate to Ring, set to YES tool has moved.
    Probe 1 - Calibrate to Ring, set to NO tool has moved.
    Probe 1 - Calibrate to Sphere, set to YES tool has moved.
    Probe 2 - Calibrate to Sphere, set to NO tool has moved.
    Probe 3 - Calibrate to Sphere, set to NO tool has moved.
    etc.
      Does it matter what order the probes are in the program? In this case, I'm running the tactile probe first and then the vision.
      I remeasured both probes as described above, but got the same results as before when measuring the same part feature both ways.
        Are you able to post your code at all? If so, use the [ CODE ] bracket at the beginning of your code and the [ / CODE ] at the end of your code. (don't put spaces in between the brackets and letters. I did for example purposes).
          The program is only for testing purposes. It measures the same ring used for setting up probes.
          PART NAME  : test of vision system using tactile
          REV NUMBER :
          SER NUMBER :
          STATS COUNT : 1
          STARTUP    =ALIGNMENT/START,RECALL:USE_PART_SETUP,LIST=YES
                      ALIGNMENT/END
                      MODE/MANUAL
                      LOADPROBE/SF_4BY10
                      TIP/TIP1, SHANKIJK=-0.1072, 0.1006, 0.9891, ANGLE=0.5665
                      MODE/MANUAL
          PLN_MAN    =FEAT/PLANE,CARTESIAN,TRIANGLE
                      THEO/<0,0,0>,<0,0,1>
                      ACTL/<-2.1283,2.9909,-8.0709>,<-0.0000083,0.0014761,0.9999989>
                      MEAS/PLANE,4
                        HIT/BASIC,NORMAL,<-1.4814,2.918,-8.0708>,<-0.0005,0,0.9999999>,<-2.2353,2.5165,-8.0702>,USE THEO=YES
                        HIT/BASIC,NORMAL,<-1.4814,2.918,-8.0708>,<-0.0005,0,0.9999999>,<-1.9858,3.4138,-8.0715>,USE THEO=YES
                        HIT/BASIC,NORMAL,<-1.4814,2.918,-8.0708>,<-0.0005,0,0.9999999>,<-1.6479,2.9957,-8.0709>,USE THEO=YES
                        HIT/BASIC,NORMAL,<-1.4814,2.918,-8.0708>,<-0.0005,0,0.9999999>,<-2.6443,3.0375,-8.071>,USE THEO=YES
                      ENDMEAS/
          CIR_MAN    =FEAT/CIRCLE,CARTESIAN,IN,LEAST_SQR
                      THEO/<-2.1493,2.9374,-8.1252>,<0,0,1>,0.7875
                      ACTL/<-2.1493,2.9375,-8.0762>,<0,0,1>,0.7875
                      MEAS/CIRCLE,4,ZPLUS
                        HIT/BASIC,NORMAL,<-1.7556,2.9422,-8.1251>,<-0.9999267,-0.0121087,0>,<-2.1644,2.544,-8.0767>,USE THEO=YES
                        HIT/BASIC,NORMAL,<-2.5423,2.9607,-8.1251>,<0.9982603,-0.0589614,0>,<-2.0585,3.3206,-8.0767>,USE THEO=YES
                        HIT/BASIC,NORMAL,<-2.2668,2.5617,-8.1252>,<0.2985523,0.9543933,0>,<-2.5367,3.0081,-8.0766>,USE THEO=YES
                        HIT/BASIC,NORMAL,<-2.0869,3.3262,-8.1253>,<-0.1585659,-0.9873484,0>,<-1.7613,2.8702,-8.075>,USE THEO=YES
                      ENDMEAS/
          A1         =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
                        ALIGNMENT/LEVEL,ZPLUS,PLN_MAN
                        ALIGNMENT/TRANS,ZAXIS,PLN_MAN
                        ALIGNMENT/TRANS,XAXIS,CIR_MAN
                        ALIGNMENT/TRANS,YAXIS,CIR_MAN
                      ALIGNMENT/END
                      MODE/DCC
                      MOVESPEED/ 50
                      PREHIT/0.05
                      RETRACT/0.05
                      MOVE/POINT,NORMAL,<0,0,0.1>
          PLN_DCC    =FEAT/PLANE,CARTESIAN,TRIANGLE
                      ACTL/<0,0,0.0012>,<0.0000115,-0.0000276,1>
                      MEAS/PLANE,4
                        HIT/BASIC,NORMAL,<-0.5,0,0>,<0,0,1>,<-0.5,0,0.0012>,USE THEO=YES
                        HIT/BASIC,NORMAL,<0,0.5,0>,<0,0,1>,<-0.0001,0.5001,0.0012>,USE THEO=YES
                        HIT/BASIC,NORMAL,<0.5,0,0>,<0,0,1>,<0.4999,0,0.0012>,USE THEO=YES
                        HIT/BASIC,NORMAL,<0,-0.5,0>,<0,0,1>,<0,-0.5,0.0012>,USE THEO=YES
                      ENDMEAS/
          CIR_DCC    =FEAT/CIRCLE,CARTESIAN,IN,LEAST_SQR
                      THEO/<0,0,-0.05>,<0,0,1>,0.788
                      ACTL/<0,-0.0003,-0.05>,<0,0,1>,0.7875
                      MEAS/CIRCLE,4,ZPLUS
                        HIT/BASIC,NORMAL,<-0.394,0,-0.05>,<1,0,0>,<-0.3938,0.0001,-0.0501>,USE THEO=YES
                        HIT/BASIC,NORMAL,<0,0.394,-0.05>,<0,-1,0>,<-0.0001,0.3934,-0.0499>,USE THEO=YES
                        HIT/BASIC,NORMAL,<0.394,0,-0.05>,<-1,0,0>,<0.3937,0.0001,-0.0499>,USE THEO=YES
                        HIT/BASIC,NORMAL,<0,-0.394,-0.05>,<0,1,0>,<0,-0.3941,-0.05>,USE THEO=YES
                      ENDMEAS/
          A2         =ALIGNMENT/START,RECALL:A1,LIST=YES
                        ALIGNMENT/LEVEL,ZPLUS,PLN_DCC
                        ALIGNMENT/TRANS,ZAXIS,PLN_DCC
                        ALIGNMENT/TRANS,XAXIS,CIR_DCC
                        ALIGNMENT/TRANS,YAXIS,CIR_DCC
                      ALIGNMENT/END
          $$ NO,
                      Tactile Inspection
          CYL1       =FEAT/CONTACT/CYLINDER/DEFAULT,CARTESIAN,IN,LEAST_SQR
                      THEO/<0,0,0>,<0,0,1>,0.788,0.1
                      ACTL/<0,0,0>,<-0.0000901,-0.0003603,0.9999999>,0.7875,0.1
                      TARG/<0,0,0>,<0,0,1>
                      START ANG=0,END ANG=0
                      ANGLE VEC=<1,0,0>
                      DIRECTION=CCW
                      SHOW FEATURE PARAMETERS=YES
                        VOID DETECTION=NO
                        REMEASURE=NO,USE THEO=NO
                        SURFACE=THICKNESS_NONE,0
                        MEASURE MODE=NOMINALS
                        RMEAS=NONE,NONE,NONE
                        CIRCULAR MOVES=STRAIGHT
                        GRAPHICAL ANALYSIS=NO
                        FEATURE LOCATOR=NO,NO,""
                      SHOW CONTACT PARAMETERS=YES
                        NUMHITS=4,NUMLEVELS=2,DEPTH=0.02,END OFFSET=0.02,PITCH=0
                        SAMPLE METHOD=SAMPLE_HITS
                        SAMPLE HITS=0,SPACER=0
                        AVOIDANCE MOVE=NO,DISTANCE=0.3937
                      SHOW HITS=NO
                      MOVE/POINT,NORMAL,<0,0,2>
          $$ NO,
                      Vision Inspection
                      LOADPROBE/VISIO_6X-1X
                      TIP/TIP1, SHANKIJK=-0.1174, 0.9871, 0.1092, ANGLE=138.5309
          CIR1       =FEAT/VISION/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                      THEO/<0,0,0>,<0,0,1>,0.788,0,0
                      ACTL/<0,-0.0002,0.0023>,<-0.0000031,-0.0014485,0.999999>,0.7875,0,0
                      ANGLE VEC=<1,0,0>
                        SURFACE=THEO_THICKNESS,0
                        PREPOSITION=NO
                        DIAGNOSTICS=NO
                      SHOW_VISION_PARAMETERS=YES
                        TYPE=AUTOMATIC HIT TARGET
                        COVERAGE=100%
                        COVERAGE ACTIVE TARGETS=4
                        MAGNIFICATION=0.7548
                        HIT TARGET COLOR=BLACK,NOMINAL COLOR=GREEN
                        HIT TARGET=EA1,0,1,NORMAL,0.0596
                        FILTER=YES,2,YES,0,2
                        EDGE=MATCHING EDGE,9,[?]->[?],-1,-1,-1,NO,Top Light=<OFF,40,OFF>,Bottom Light=<OFF,40,ON>,Ring Light=<OFF,0.8,OFF>
                        FOCUS=YES,FULL,0.1,YES,2,NONE,NO
          PNT1       =FEAT/VISION/SURFACE POINT/DEFAULT,CARTESIAN
                      THEO/<0.01,0.55,0>,<0,0,1>
                      ACTL/<0.01,0.55,0.0124>,<-0.0000031,-0.0014485,0.999999>
                      TARG/<0.01,0.55,0>,<0,0,1>
                      SNAP=NO
                        PREPOSITION=NO
                        DIAGNOSTICS=NO
                        HIT TARGET=SA1,NO,0.25,0.25
                        FOCUS=AUTO,YES,0.1968504,,NONE,-1,-1,-1,NO,Top Light=<OFF,40,OFF>,Bottom Light=<OFF,40,OFF>,Ring Light=<OFF,95,ON>
          PNT2       =FEAT/VISION/SURFACE POINT/DEFAULT,CARTESIAN
                      THEO/<0.01,-0.55,0>,<0,0,1>
                      ACTL/<0.01,-0.55,0.0128>,<-0.0000031,-0.0014485,0.999999>
                      TARG/<0.01,-0.55,0>,<0,0,1>
                      SNAP=NO
                        PREPOSITION=NO
                        DIAGNOSTICS=NO
            Yeah they won't measure "exactly the same". Vision is a different beast. The -0.0002 is common because vision is picking up many many more hits that your 8 hit cylinder. In addition, to the Z height variation, your vision probe has focus enabled, so that is the depth that it focused on to obtain the diameter it did.
              Thanks. That's the answer I was looking for.
              So, this means I should re-align the part after switching to vision?
                No. I don't think you should need to. Your cylinder is only 0.1 long. You could try, but I doubt it will do anything.