whats the standard operating air pressure for a cmm? our compressor has a leak at one of the regulartors and the air pressure is at 20psi.
PSI
Collapse
X
-
i paged maintenance, four guys show up within minutes (problably bored from reading the paper). they drained the commpressor and cleaned the regulators. there is a huge puddle on the floor from all the water in the compressor and the glass on the regulators where just caked with yellow guuk and are now clean. running at 60psi, don't recall what it was before the problem but i have a call into tech sup to see what the normal operating range is for our machine.
Comment
-
Be careful now, your air must be CLEAN AND DRY too! Most manufacturers specify you also buy an air dryer. The reason is that your air bearings have orfices that are typically no bigger than 0.015". If you have water or dirt in the line, these bearings can easily clog. Best case, they cause your squareness to change. Worst case, they drag and get destroyed. Putting new bearings in is probably almost as expensive as buying a new machine!
The FilterRegulator (FR) unit that your CMM vendor supplies can only handle "nominal" amounts of dirt and water. If I were you, have maintenance instal filters right after the compressor, as well as an air dryer (the CMM vendor can tell you what the SCFM is, for your machine probably between 10 and 20). You will run your CMM trouble free for much longer.
Also, ask your techs whether they can make you a water trap: this is a piece of dead end line that is close to the lowest point of the circuit with a ball valve. Every morning, open the ball valve and let the water out. Works great and again prevents water to enter the air bearings.
If moisture gets into your machine, you are on your way to catastrophic failure. I suggest you take this very serious!
Jan.***************************
PC-DMIS/NC 2010MR3; 15 December 2010; running on 18 machine tools.
Romer Infinite; PC-DMIS 2010 MR3; 15 December 2010.
Comment
-
Originally posted by Jan d.Also, ask your techs whether they can make you a water trap: this is a piece of dead end line that is close to the lowest point of the circuit with a ball valve. Every morning, open the ball valve and let the water out. Works great and again prevents water to enter the air bearings.
If moisture gets into your machine, you are on your way to catastrophic failure. I suggest you take this very serious!
Jan.
Comment
-
Originally posted by Jan d.Be careful now, your air must be CLEAN AND DRY too! Most manufacturers specify you also buy an air dryer. The reason is that your air bearings have orfices that are typically no bigger than 0.015". If you have water or dirt in the line, these bearings can easily clog. Best case, they cause your squareness to change. Worst case, they drag and get destroyed. Putting new bearings in is probably almost as expensive as buying a new machine!
Originally posted by Jan d.Also, ask your techs whether they can make you a water trap: this is a piece of dead end line that is close to the lowest point of the circuit with a ball valve. Every morning, open the ball valve and let the water out. Works great and again prevents water to enter the air bearings.
Originally posted by Jan d.If moisture gets into your machine, you are on your way to catastrophic failure. I suggest you take this very serious!
Jan.When in doubt, post code. A second set of eyes might see something you missed.
sigpic
Comment
-
I am running MicroXcel PFx and the pressure must be set at 55 +/-1 psi.
Be careful if the pressure is too low the air bearings can land on the rails and damage them.
Originally posted by uber userwhats the standard operating air pressure for a cmm? our compressor has a leak at one of the regulartors and the air pressure is at 20psi.N.
Comment
-
Machine Problem
I have two Browne & Sharpe DEA machines. We have air dryers and a catch can so to speak before it gets to our room. One of my old CMM Operators was working second and two of the airbearing on the 'Z' rails locked up and shaved a chunk from top to bottom from the 'Z' rail. It was about a ten thousand dollar problem to fix. My plant takes us serious now when we need things. I haven't had any problems since we booted her out of here. After it stripped the z rail clean she kept running it the rest of the night. They take us serious when we tell them that they can't just put anyone running a cmm either.REVERSE, REVERSE, REVERSE
Comment
-
It's amazing sometimes. Management thinks if they can walk,talk and breath then they can run a CMM. I've had people i know could not find their way to work without directions and i was suppose to train them. Amazingsigpic.....Its called golf because all the other 4 letter words were taken
Comment
Related Topics
Collapse
-
Hello, I'm wondering at what PSI do you guys run your machines at.
currently getting 80 PSI from our air supply and the machine is set at 65 PSI?...-
Channel: PC-DMIS for CMMs
07-01-2020, 03:43 PM -
-
I have an air pressure fault or lube fault error. I have air supply to the CMM but I can't get the machine to start. What's next? Currently running M...
-
Channel: PC-DMIS for CMMs
08-20-2007, 10:14 AM -
-
by BIGWIG7Hey fellas,
I'm having a hard time with the machine randomly losing air pressure.
I know for a fact that the CMM is being provided...-
Channel: PC-DMIS for CMMs
01-15-2015, 10:03 AM -
-
Can anyone tell me what the PSI should be for a Global Image 12-15-10, I think it is 80 but not sure.
-
Channel: PC-DMIS for CMMs
12-05-2012, 03:25 PM -
-
by Navin R.I'm at a new job-the CMM there-a Global Advantage. We've got manuals for the software, but nothing about the CMM itself. Can anyone tell me what is...
-
Channel: PC-DMIS for CMMs
05-14-2008, 07:17 PM -
Comment