I have a 4-Probe Changer, PCDMIS 3.5 v2, is there a method to Calibrate all of the Probes one-after-another in the same operation?
Probe Calibration
After you define you probe changer. You can write a program that has all 4 of your probes. Here is what mine looks like. (Sp600) changer
PART NAME : AUTOCALIBRATE
REV NUMBER :
SER NUMBER :
STATS COUNT : 1
STARTUP =ALIGNMENT/START,RECALL:STARTUP, LIST= YES
ALIGNMENT/END
MODE/MANUAL
DISPLAYPRECISION/4
TEMPCOMP/ORIGIN=0,0,0,Material Coeff=0.000006389,Reference Temp=68
,Hi Threshold=104,Lo Threshold=50,Sensor num=1
,X Axis Temp=71.1893,Y Axis Temp=70.9822,Z Axis Temp=71.1783,Part Temp=70.2337
OPTIONPROBE/MAXFORCE=0.3, LOWFORCE=0.05, UPPERFORCE=0.15, TRIGGERFORCE=0.1,
PROBEACCURACY=0.1, POSACCURACY=0.1, #RETURNDATA=7,
RETURNSPEED=1, SCANPNTDENSITY=2,
SCANACCELERATION=10, SCANOFFSETFORCE=0.1,
PROBINGMODE=DFL, MANFINEPROBING=NO
FORMAT/TEXT,OPTIONS, ,HEADINGS,SYMBOLS, ;NOM,MEAS,TOL,DEV,OUTTOL, ,
PREHIT/ 0.1
RETRACT/ 0.1
CHECK/ 0.1,1
MOVESPEED/ 40
TOUCHSPEED/ 0.5
LOADPROBE/PROBE1
AUTOCALIBRATE/PROBE, PARAMETER_SET=PROBE1, QUALTOOL_MOVED=YES,
SHOW_SUMMARY=NO, OVERWRITE_RESULTSFILE=YES
LOADPROBE/PROBE2
AUTOCALIBRATE/PROBE, PARAMETER_SET=PROBE2, QUALTOOL_MOVED=NO,
LOADPROBE/PROBE3
AUTOCALIBRATE/PROBE, PARAMETER_SET=probe3, QUALTOOL_MOVED=NO,
LOADPROBE/PROBE4
AUTOCALIBRATE/PROBE, PARAMETER_SET=probe4, QUALTOOL_MOVED=NO,
SHOW_SUMMARY=NO, OVERWRITE_RESULTSFILE=YESTime for the Trolls to leave.
by Mike N.Anyone have advice for writing an automated routine to calibrate all the probes in a rack?
When using a "Auto Calibrate" command...-
04-24-2006, 08:30 AM -
by smuchDoes anyone know how to automate the calibration of all probes and positions of probes in PC-DMIS? Yes, its not too difficult to
click on all...-
12-19-2007, 06:14 PM -
by B. JacobsOK guy's. Here is an easy one(I hope).
Can someone tell me an easy way of writing a program that will calibrate all the probes I have in my probe...-
09-07-2006, 03:03 PM -
by DarrollhGood morning,
I've been asked to put in a probe calibration program to calibrate probes at the beginning of the day. I've never had to do this before....-
10-23-2017, 11:14 AM -