We have been tasked with providing some form analysis for some repair sleeves that go into parts we refurbish for the Army.
In order to get ready for that, I wanted to determine how well a TP2 and TP200 compared when probing a ring gage.
I knew from the design of the trigger mechanism of the TP2 that I would see lobing. What I did NOT expect was to see lobing on the TP200.
I thought the TP200 was a transducer and lobing was not an issue with the design. It seems that assumption is incorrect.
We measured 360 points both with the TP2 and TP200 and we are seeing essentially the same data. Diameter variation and point deviation correlation is out to the 5th decimal place.
We have turned the ring 90 degrees to see if the lobing is in the ring or the probe. Program is running as I type this. The ring I have is a Bowers 46mm bore setting ring. It is the best class ring I can get my hands on at this time.
Any of you have similar experiences?
Thanks!
Hilton
In order to get ready for that, I wanted to determine how well a TP2 and TP200 compared when probing a ring gage.
I knew from the design of the trigger mechanism of the TP2 that I would see lobing. What I did NOT expect was to see lobing on the TP200.
I thought the TP200 was a transducer and lobing was not an issue with the design. It seems that assumption is incorrect.
We measured 360 points both with the TP2 and TP200 and we are seeing essentially the same data. Diameter variation and point deviation correlation is out to the 5th decimal place.
We have turned the ring 90 degrees to see if the lobing is in the ring or the probe. Program is running as I type this. The ring I have is a Bowers 46mm bore setting ring. It is the best class ring I can get my hands on at this time.
Any of you have similar experiences?
Thanks!
Hilton
Comment