    Can anyone tell me how to change the calibration program to make the probe head go to the upper center of the table for safe rotation before it goes to another calibration angle?
    I am calibrating a scanning head and it takes a very long time to calibrate. As it stands I have to baby sit it so it doesn't wreck.
    Eric

  • #2
    In the probe utilities window under "SETUP" you can change the clearance distance from the tool.
    I used to be high on life but I built up a tolerance.

    Brown & Sharpe Global Advantage
    PCDMIS CAD++ v2011mr2
    PH10MQ/SP600M


    sigpic

    Comment


    • #3
      open the probe utilities window and select "set-up"

      change the "Clearance distance along the qualification tool shank vector:"

      mine is 3.5 (inches) and works fine

      Comment


      • #4
        what they said. But you could also just do one angle at a time. I have the SP25
        sigpicSummer Time. Gotta Love it!