  • Strange Calibration Issue

    I just calibrated a tip (8x50mm) in the A0B0 and everything looks great. I had .00006 STD.

    All my other tips are longer than 100mm and every tip is being reported as exceeding limit. All tips are being calibrated at 1% touch speed.

    Can anyone tell me what you are setting your parameters at for tips longer than 100mm. My longest is at 150mm. Is it a deflection issue? Should I adjust my trigger force?
    I used to be high on life but I built up a tolerance.

    Brown & Sharpe Global Advantage
    PCDMIS CAD++ v2011mr2
    PH10MQ/SP600M


    sigpic

  • #2
    On my XCEL12-20-10 I use a touch speed of .75 and most of my probes are longer than 100mm.
    I have a tp200 also on this machine.
    on the Global I have a SP600 and the touch speed is .5 on it.
    have you ever done a lower matrix calibration on your SP600. It is located in the probe calibration window. This might help.
    When you do it use your 8x50mm as it is the most rigid one that you probably have.
    Time for the Trolls to leave.

    Comment


    • #3
      Where are you seeing this lower matrix calibration?
      I used to be high on life but I built up a tolerance.

      Brown & Sharpe Global Advantage
      PCDMIS CAD++ v2011mr2
      PH10MQ/SP600M


      sigpic

      Comment


      • #4
        With my SP600M im using a 5mm/sec Touch Speed. 25 hits 4 levels

        For StDev I have my warning setup at .01

        Probe Options:
        Max Force 0.3N
        Low Force 0.06N
        Upper Force 0.18N
        Trigger Force 0.12N
        Mr. Comment
        SCIROCCO-NT 13-20-10
        B3C-LC Controller (Leitz Protocol), SP600M, TP200
        PCDMIS CAD++ v4.3 MR1(Build: 12/11/08)
        sigpic

        Comment


        • #5
          Oh nevermind I found it, what does this do? Never used it.
          I used to be high on life but I built up a tolerance.

          Brown & Sharpe Global Advantage
          PCDMIS CAD++ v2011mr2
          PH10MQ/SP600M


          sigpic

          Comment


          • #6
            Originally posted by Underspec View Post
            I just calibrated a tip (8x50mm) in the A0B0 and everything looks great. I had .00006 STD.

            All my other tips are longer than 100mm and every tip is being reported as exceeding limit. All tips are being calibrated at 1% touch speed.

            Can anyone tell me what you are setting your parameters at for tips longer than 100mm. My longest is at 150mm. Is it a deflection issue? Should I adjust my trigger force?
            'exceeding limit' ??? diameter or stdev?

            Did you check to make sure they are clean and tight?

            Comment


            • #7
              Low level matrix is a calibration to compenate for errors within the analog probe itself.
              Mr. Comment
              SCIROCCO-NT 13-20-10
              B3C-LC Controller (Leitz Protocol), SP600M, TP200
              PCDMIS CAD++ v4.3 MR1(Build: 12/11/08)
              sigpic

              Comment


              • #8
                Just did a lower level matrix calibration using 8x50mm and says my cal diameter exceeds limit
                I used to be high on life but I built up a tolerance.

                Brown & Sharpe Global Advantage
                PCDMIS CAD++ v2011mr2
                PH10MQ/SP600M


                sigpic

                Comment


                • #9
                  Originally posted by Underspec View Post
                  Just did a lower level matrix calibration using 8x50mm and says my cal diameter exceeds limit
                  In that case, double check your probe build to ensure that the tip diameter is as it should be. Also, you may want to check your cal. sphere definition to ensure it is defined properly for those probes.

                  Comment


                  • #10
                    Originally posted by Goodluck View Post
                    In that case, double check your probe build to ensure that the tip diameter is as it should be. Also, you may want to check your cal. sphere definition to ensure it is defined properly for those probes.
                    Don't forget to check your scroll-lock key as well!
                    sigpic
                    Originally posted by AndersI
                    I've got one from September 2006 (bug ticket) which has finally been fixed in 2013.

                    Comment


                    • #11
                      Originally posted by Goodluck View Post
                      In that case, double check your probe build to ensure that the tip diameter is as it should be. Also, you may want to check your cal. sphere definition to ensure it is defined properly for those probes.

                      The tips are built correctly as far as I know. And the Cal sphere dia. is set to exactly as what is on the cal sphere.
                      I used to be high on life but I built up a tolerance.

                      Brown & Sharpe Global Advantage
                      PCDMIS CAD++ v2011mr2
                      PH10MQ/SP600M


                      sigpic

                      Comment


                      • #12
                        My shortest tip is calibrating ok. But all my other longer tips are exceeding limits. Any thoughts?
                        I used to be high on life but I built up a tolerance.

                        Brown & Sharpe Global Advantage
                        PCDMIS CAD++ v2011mr2
                        PH10MQ/SP600M


                        sigpic

                        Comment


                        • #13
                          Originally posted by Underspec View Post
                          My shortest tip is calibrating ok. But all my other longer tips are exceeding limits. Any thoughts?
                          nope - sorry.

                          wait - yes - I think this is a user editable setting. Does it tell you what diameter it is getting for each of the tips? Are they the same and your short tip is set with a greater tolerance than the others?

                          Comment


                          • #14
                            Are they carbon or steel...the longer ones
                            sigpic.....Its called golf because all the other 4 letter words were taken

                            Comment


                            • #15
                              Tips are carbon and ceramic. Cal tool is ceramic.
                              I used to be high on life but I built up a tolerance.

                              Brown & Sharpe Global Advantage
                              PCDMIS CAD++ v2011mr2
                              PH10MQ/SP600M


                              sigpic

                              Comment

                              Related Topics

                              Collapse

                              • ag1979
                                Trigger Force
                                by ag1979
                                Has anyone used, or does anyone use a gram gauge to set the proper triggering speed. And does anyone know a good way to figure out the optimum trigger...
                                04-21-2006, 11:25 AM
                              • Tested
                                SP600 pause
                                by Tested
                                We have a SP600 and when used to take points it pauses after each touch.
                                I thought that was just the way the SP600 worked but we had Hexagon in...
                                09-09-2016, 02:57 PM
                              • grandpaal56
                                Tp7 touch speeds
                                by grandpaal56
                                I'm just curious. We use primarily TP7 probes. I know they don't like to be babied when it comes to touch speeds, so I'm wondering what others who use...
                                11-19-2011, 10:44 AM
                              • zach19564
                                lower trigger force
                                by zach19564
                                im tring to do the calibration on a new probe. when dcc trys to take its first point it comes up with an error message "lower tigger not reached"....
                                03-10-2013, 02:32 PM
                              • Jackarroo
                                Probing Force Help!
                                by Jackarroo
                                I searched, but couldn't find an answer..... Is there some sort of standard for how much to reduce probing force, as you step down in size of stylus tips?...
                                02-15-2018, 06:04 PM
                              Working...
                              X