    cmm calibration just finished and shows total volumetric performance whatever at .0006" Is this acceptible, he says no adjustment is needed, yet i have to inspect parts with ledss than that as a tolerance. even at .001" tolerance, this can be out due to this deviation? im not sure
    That is a lot. B & S specs to .002. Mine runs at .0002 volumetric. But that is different than Repeatability. Mine repeats to .00003 in a 2.0000 inch ring gage.
    G
      You have to look at the factory specs for your machine. You can not expect your machine to be any better than the factory said it could be. If you are checking things with a total tolerance of 0.001", then your machine (factory specs AND calibration results) SHOULD be 0.0001" or less (10% of the total tolerance).
        What type of machine, do you use TP2, TP20,TP200
        Your volumetric is like an measuring envelope for the variation your machine measured during the B89 certification. Did the use the ball bar.
        .001 is to tight for my machine for me to trust the accuracy & repeatability.
          phto t and tp20 extended force, i meant to put +/-.001" tolerance
            If it was me I'd tell the accuracy is not there...my .02
              You can NOT expect ANY calibration tech (no matter WHERE they are from) to make the machine better than the factory says it is going to be. If the factory specs on his machine (the machine, not his set-up items, I don't care if he has a PH9 and a TP2 with a 75mm long tip or an SP600 with a 5mm long tip) is total volumetric of 0.0008" and the calibration tech shows it at 0.0007", then he IS done. It does not matter if YOU want it better, it is already showing better than the factory said it could be, period. That is the end of it.
              Sure, we would ALL like to see calibration values of 0.0001" or less, but if the machine is NOT made for it, then you CAN NOT expect to see numbers that low.
                Some days its not even worth chewing through the restraints.
                http://www.ericrogers.org/photos/tra...s/DSCN3106.jpg
                  In a similar line of thought... go back and double check the cal. numbers. Are you sure they are in inches? Last time mine was done, at first glance the numbers looked huge but then I realized he calibrated it in mm.