Multiple Qual Spheres
  • Multiple Qual Spheres
    Is it posible to use multiple qual spheres....
    1. calibrate a star probe on a qual sphere vector 0,1,0
    2 calibrate a single probe on a diffrent sphere vector 0,0,1
    zeroed on the qual sphere with star probe....
    checked location of sphere with 5x50...off by .045 3d...
    requaled 5x50 on 0,1,0 vector qual sphere and ran...within .0004 3d
    Yes, it is possible. Follow my linked info exactly and you should have no problem. That is if you get through all "Jimi-isms". Any questions Tim jus write back.
    Go...
    HERE
      2 qual sphere 2nd question
      OK.......
      are the parameter sets a necesity......
      1. I calibrated (tp20) m5x50 a0b0.a90b0,a90b180 on qual sphere1 (0,0,1)
      2. I calibrated m5x50 a0b0 on qual sphere2 (0,-1,0)
      3. I should now calibrate all except 1 tip of star probe on 0,0,1 sphere, then calibrate last tip of star on 0,-1,0 sphere??
        Originally posted by Boomer View Post
        OK.......
        Sorry, I was going to report back to you that the parameter sets are not necessary but got busy when I first responded.
        Answer to #3...That sounds correct. Qual all tips(including 0,0) that you can on #1 sphere. Qual 0,0 on #2 sphere answering "yes" it has moved and that is what creates the relationship between #1 sphere and #2 sphere. Now calibrate all tips on #2 sphere you need to answering "no" for the rmainder of the tips.
          Thanks
          Jimi......
          Worked like a charm....thank you