Hole diameter not even close to accurate
  • Hole diameter not even close to accurate
    Hello all,
    I'm using an auto feature picked from the cad for an auto circle. Nominal diameter is 10, using 3 sample hits spaced 2 mm out, 4 hits for the feature, depth of .3mm (sheet metal).
    The measurements I'm getting from it are 6.5mm for diameter. I can't figure out why. It's the only hole in this same surface that is causing issue, all the others are measuring fine.
    I checked calibration for this probe angle, got a STDDEV of .001
    I'm not picking up any significant bur in the hole, or at least none visible to cause a 3.5mm outtol.
    I've tried increasing hits, change depth, change spacer, change start/end angle: nothing. Cant get it to budge.
    Recalibrated, nothing changed.
    The weird thing I noticed when I was messing with the start/end angle:
    Instead of measuring 0-360, as I would for every hole (because my probe is almost perfectly perpendicular to the feature's surface), I tried measuring each quadrant. so 0-90, 90-180, 180-270, 270-360 trying to see if maybe I have a severely oblong hole. What was really strange is when I measured 0-90, It would take the 3 sample hits in the correct quadrant, but would measure the 4 feature hits in the 180-270 quadrant, rather than in the 0-90 as it should. Regardless, all the measurements fell in the mid-6mm zone, so it still proved nothing.
    Anyone have some thoughts on this? If you will teach me how to post my code, I will do so.
    type in [code] then copy & paste your code, then type in [/code] after the code.
    As for the size issue, what size probe are you using? Maybe Pcd is confused and is comping like it is an 'outside' instead of an 'inside' circle.
    • bigblue97mustang
      bigblue97mustang commented
      2mm ruby in a 10mm hole. Code coming up.
    I should also note, I tried a completely different probe build on a different tp20, and got the same measurement.
      tartup =alignment/start,recall:use_part_setup,list=yes
      alignment/end
      mode/manual
      movespeed/ 250
      touchspeed/ 5
      fly/on
      format/text,options, ,headings,symbols, ;nom,tol,meas,dev,outtol, ,
      loadprobe/100_tp20_sf_30x2
      tip/t1a0b0, shankijk=0, 0, 1, angle=0
      move/point,normal,<753.248,858.19,-58.339>
      move/point,normal,<1929.927,821.999,90.199>
      pnt1 =feat/contact/surface point/default,cartesian
      theo/<1933.239,736.6,89.371>,<0,1,0>
      actl/<-326.029,-1180.431,-1974.842>,<-0.0014123,0.999999,-0.000208>
      targ/<1933.239,736.6,89.371>,<0,1,0>
      show feature parameters=no
      show contact parameters=yes
      sample method=sample_hits
      sample hits=3,spacer=2
      avoidance move=both,distance=5
      show hits=no
      sltr1 =feat/contact/round slot/default,cartesian,in
      theo/<1815,801,43>,<0,1,0>,<-1,0,-0.0000101>,12,14
      actl/<-445.564,-1116.611,-2021.14>,<0.0055959,0.9999113,0.0120861>,<-0.9999002,0.0054383,0.0130358>,11.938,13.924
      targ/<1815,801,43>,<0,1,0>,<-1,0,-0.0000101>
      meas angle=90
      numhits=6,depth=0.3
      find hole=disabled,onerror=no,read pos=no
      show hits=no
      move/point,normal,<1758.825,829.906,105.458>
      move/point,normal,<820.721,831.289,104.545>
      pnt2 =feat/contact/surface point/default,cartesian
      theo/<751.368,736.2,90.193>,<0,1,0>
      actl/<-1507.899,-1182.5,-1972.896>,<-0.0014123,0.999999,-0.000208>
      targ/<751.368,736.2,90.193>,<0,1,0>
      show hits=no
      move/point,normal,<752.305,811.273,91.032>
      cir1 =feat/contact/circle/default,cartesian,in,least_sqr
      theo/<750,801,43>,<0,1,0>,12
      actl/<-1509.413,-1117.224,-2020.115>,<0.0008259,0.9999954,0.0029299>,11.925
      targ/<750,801,43>,<0,1,0>
      start ang=0,end ang=360
      angle vec=<-1,0,0>
      direction=ccw
      numhits=4,depth=0.3,pitch=0
      show hits=no
      move/point,normal,<747.67,837.924,-37.666>
      pnt3 =feat/contact/surface point/default,cartesian
      theo/<748.278,736.2,-57.186>,<0,1,0>
      actl/<-1511.121,-1182.535,-2120.29>,<-0.0014123,0.999999,-0.000208>
      targ/<748.278,736.2,-57.186>,<0,1,0>
      show hits=no
      comment/oper,no,full screen=no,auto-continue=no,
      dcc
      mode/dcc
      a1 =alignment/start,recall:startup,list=yes
      alignment/iterate
      pnt target rad=0.25,start label=,fixture tol=0.05,error label=
      meas all feat=always,max iterations=99,level axis=yaxis,rotate axis=zaxis,origin axis=xaxis
      level=pnt1,pnt2,pnt3,,
      rotate=cir1,sltr1,,
      origin=cir1,,
      alignment/end
      prehit/4
      8-1-3 =feat/contact/circle/default,cartesian,in,least_sqr
      theo/<1195,778.223,69.592>,<0,0.1391731,0.9902681>,10.0 01
      actl/<1193.868,778.469,71.119>,<-0.003003,0.1540237,0.9880626>,6.137
      targ/<1195,778.223,69.592>,<0,0.1391731,0.9902681>
      start ang=0,end ang=-360
      angle vec=<1,0,0>
      direction=cw
      show hits=no
      dim dim 8-1-3= location of circle 8-1-3 units=mm ,$
      graph=off text=off mult=10.00 output=both half angle=no
      ax nominal +tol -tol meas dev outtol
      x 1195.000 1.000 1.000 1193.868 -1.131 0.131 <--------
      y 778.223 1.000 1.000 778.469 0.246 0.000 -----#---
      z 69.592 1.000 1.000 71.119 1.526 0.526 -------->
      d 10.001 0.100 0.100 6.137 -3.864 3.764 <--------
      end of dimension dim 8-1-3
        Add the form of the circle to the dimension, is it round?
        sigpicHave a homebrew
        • ctw00
          ctw00 commented
          bigblue97mustang, I see that the X deviation is 1.13 and the Y is .247, is that typical? I don't think the size issue is the result of cosine error as it shouldn't have that much error, but you might try changing the Find Hole setting to center. It will take 3 points inside the circle and roughly center the probe before checking the hole.
        • bigblue97mustang
          I think I tried a find center as well, but can't say for absolute certainty. And these are pre-trial parts, so a 1mm deviation from center is somewhat normal. Especially since all the features around that circle are deviated in the same direction.
          They still have plenty of jig tuning let to do! HAHA
        • bigblue97mustang
          @ctw00:
          No change with find hole turned on
          7_11       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                      THEO/<1195,778.223,69.592>,<0,0.1391731,0.9902681>,10
                      ACTL/<1193.869,778.491,71.127>,<-0.0013962,0.1538959,0.9880861>,6.138
                      TARG/<1195,778.223,69.592>,<0,0.1391731,0.9902681>
                      START ANG=0,END ANG=360
                      ANGLE VEC=<1,0,0>
                      DIRECTION=CCW
                      SHOW FEATURE PARAMETERS=YES
                        VOID DETECTION=YES
                        REMEASURE=NO
                        SURFACE=THEO_THICKNESS,1.6
                        MEASURE MODE=NOMINALS
                        RMEAS=NONE,NONE,NONE
                        AUTO WRIST=NO
                        CIRCULAR MOVES=STRAIGHT
                        GRAPHICAL ANALYSIS=NO
                        FEATURE LOCATOR=NO,NO,""
                      SHOW CONTACT PARAMETERS=YES
                        NUMHITS=4,DEPTH=0.3,PITCH=0
                        SAMPLE METHOD=SAMPLE_HITS
                        SAMPLE HITS=3,SPACER=2
                        AVOIDANCE MOVE=BOTH,DISTANCE=5
                        FIND HOLE=CENTER,ONERROR=NO,READ POS=NO
                      SHOW HITS=NO
          DIM TEST= LOCATION OF CIRCLE 7_11  UNITS=MM ,$
          GRAPH=OFF  TEXT=OFF  MULT=10.00  OUTPUT=BOTH  HALF ANGLE=NO
          AX    NOMINAL       +TOL       -TOL       MEAS        DEV     OUTTOL
          X     1195.000      1.000      1.000   1193.869     -1.131      0.131 <--------
          Y      778.223      1.000      1.000    778.491      0.269      0.000 -----#---
          Z       69.592      1.000      1.000     71.127      1.535      0.535 -------->
          D       10.000      0.100      0.100      6.138     -3.862      3.762 <--------
          END OF DIMENSION TEST
        Fu
        Time for the Trolls to leave.
        • bigblue97mustang
          Thats the wrong circle. That was a circle used in my alignment. The circle in question is further down.
        May be stupid but...... copy " another " hole that " worked " , change the nom's to the same location as your " troublesome " hole and run.... You get the same measurement ? Sometimes features are just corrupt....
        • bigblue97mustang
          I'm going to absolutely hate it if that works. At this point I was trying to prepare as strong of a case as I could that my CAD was causing the issue. I won't have enough time to try it out today, but I'll give it a shot first thing in the morning.
        Change the name 8-1-3 to 8_1_3.
        Probably not this, but this is the kind of weird **** stuff which can happen using math symbols in feature names.
        • bigblue97mustang
          I will change to underscore and see if that improves it.
        • bigblue97mustang
          No luck:
                      ACTL/<1193.869,778.49,71.125>,<-0.0022674,0.1539609,0.9880743>,6.139
                      ANGLE VEC=<1,0,0>
                      DIRECTION=CCW
                        VOID DETECTION=YES
                        REMEASURE=NO
                        AUTO WRIST=NO
                      SHOW HITS=NO
          Y      778.223      1.000      1.000    778.490      0.267      0.000 -----#---
          Z       69.592      1.000      1.000     71.125      1.533      0.533 -------->
          D       10.000      0.100      0.100      6.139     -3.861      3.761 <--------
        The only thing that should be in a feature name is letters, numbers, and the underscore. Nothing else.
        Also, you're using a 2mm probe, and measuring a hole 4mm undersize. That is what you would expect if it is calculating in outie from an innie. Obviously it runs through the feature, so it is measuring properly.
        What about the stupid "Void Detection" toggle (under the End Angle)? Is that toggled off?
        • VinniUSMC
          VinniUSMC commented
          bigblue97mustang It does the opposite of good things, lol
          Stupid "enhancement" to have it defaulted to ON.
        • bigblue97mustang
          Gah, the opposite of good things? That's exactly what I don't need! HAHA
        • bigblue97mustang
          Welp, I hate it, cause I have no idea wtf it's doing differently but, here's without void detection.....
          7-11       =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
                      ACTL/<1193.871,778.486,71.122>,<-0.0029247,0.154209,0.9880339>,9.927
                      ANGLE VEC=<1,0,0>
                      DIRECTION=CCW
                        VOID DETECTION=NO
                        REMEASURE=NO
                        AUTO WRIST=NO
                      SHOW HITS=NO
          DIM TEST= LOCATION OF CIRCLE 7-11  UNITS=MM ,$
          X     1195.000      1.000      1.000   1193.871     -1.129      0.129 <--------
          Y      778.223      1.000      1.000    778.486      0.263      0.000 -----#---
          Z       69.592      1.000      1.000     71.122      1.530      0.530 -------->
          D       10.000      0.100      0.100      9.927     -0.073      0.000 -#-------
        As recommended by PCDMIS,
        Feature names
        Dimension names
        Only Letters, Numbers, Underscore
        You might get away with special characters, but you are playing with fire. PCDMIS is like a hibernating bear. Don’t disturb it, because it can become ornery and will bite you in the buttocks.
        • bigblue97mustang
        • bigblue97mustang
          No change =(
                      ANGLE VEC=<1,0,0>
                      DIRECTION=CCW
                        VOID DETECTION=YES
                        REMEASURE=NO
                        AUTO WRIST=NO
                      SHOW HITS=NO
        What size do you get when you check it another way i.e. with gage pins or an ID mic?
        I'd delete it and try again. Enough effort wasted on fixing this one already ya think?
        • bigblue97mustang
          Trust me, I tried..... Even went as far as mirroring the left-hand twin of this cad and still no luck.
        8-1-3 =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR THEO/<1195,778.223,69.592>,<0,0.1391731,0.9902681>,10.0 01
        ACTL/<1193.866,778.47,71.121>,<-0.0029722,0.1540079,0.9880651>,6.138
        Looks correct to me, Let's see 8 - 1 - 3 = 4 , then take the 4 off of THEO 10, so 10 - 4 = 6, The result is 6.138, that's pretty darn close if you ask me. Looks Good!!!!
        • KIRBSTER269
          KIRBSTER269 commented
          Not really, DMIS is set up for you to math scripting inside the Editor, So when you use minus, times, addition, etc. Well when you start marking it up that way, Don't be surprised when it does.
        • bigblue97mustang
          Fair enough. I never knew there was an issue with using those in names. Maybe I just got lucky all this time.
        • bigblue97mustang
          No dice on the name change.
          7_11       =feat/contact/circle/default,cartesian,in,least_sqr
                      actl/<1193.869,778.49,71.125>,<-0.0022674,0.1539609,0.9880743>,6.139
                      angle vec=<1,0,0>
                      direction=ccw
                      show feature parameters=yes
                        void detection=yes
                        remeasure=no
                        surface=theo_thickness,1.6
                        measure mode=nominals
                        rmeas=none,none,none
                        auto wrist=no
                        circular moves=straight
                        graphical analysis=no
                        feature locator=no,no,""
                      show hits=no
          dim test= location of circle 7_11  units=mm ,$
          x     1195.000      1.000      1.000   1193.869     -1.131      0.131 <--------
          y      778.223      1.000      1.000    778.490      0.267      0.000 -----#---
          z       69.592      1.000      1.000     71.125      1.533      0.533 -------->
          d       10.000      0.100      0.100      6.139     -3.861      3.761 <--------
          end of dimension test
        bigblue97mustang
        could you change show hits=no to show hits = yes?
        And post that section of code.
        Maybe there is something there that is messing with the measurements. Bad hits, bad vectors.
        • bigblue97mustang
          I sure can. I'm working on some measurement requests currently, but I'll get this part set up again after.
        • dph51
          dph51 commented
          Did you try dimensioning the diameter again? I mean inserting a new dimension statement. Sometimes PCDMIS just gets stuck and corrupts a line a code. Inserting a new line sometimes works.
          It will turn your hair gray and sour your personality if you’re not careful. Good luck.
        HOW ABOUT CHANGING YOUR VECTORS FROM
                               theo/<1195,778.223,69.592>,<[COLOR=#c0392b]0,0.1391731,0.9902681[/COLOR]>,10
                               targ/<1195,778.223,69.592>,<[COLOR=#c0392b]0,0.1391731,0.9902681[/COLOR]>
        TO THIS
                               theo/<1195,778.223,69.592>,<[COLOR=#c0392b]0,0,1[/COLOR]>,10
                               targ/<1195,778.223,69.592>,<[COLOR=#c0392b]0,0,1[/COLOR]>
        • bigblue97mustang
          No luck.
                      THEO/<1195,778.223,69.592>,<0,0,1>,10
                      ACTL/<1193.874,778.489,71.125>,<-0.0022085,0.15415,0.988045>,6.142
                      TARG/<1195,778.223,69.592>,<0,0,1>
                      ANGLE VEC=<1,0,0>
                      DIRECTION=CCW
                        VOID DETECTION=YES
                        REMEASURE=NO
                        AUTO WRIST=NO
                      SHOW HITS=NO
          X     1195.000      1.000      1.000   1193.874     -1.126      0.126 <--------
          Y      778.223      1.000      1.000    778.489      0.266      0.000 -----#---
          D       10.000      0.100      0.100      6.142     -3.858      3.758 <--------
        • KIRBSTER269
          Did you re-run it?
        • bigblue97mustang
          Yeah, I did. Tried to dimension IJK in the test dimension, but didn't see an option for it.