    Am measuring a chamfer diameter and angle. Per my customer's spec, the tolerance of the angle varies and is driven by the measured size of the cone. How can I make PCD look at the diameter and then automatically change the tolerance it uses for the chamfer angle's dimensional output?
    Cone nominal Ø 0.6880-0.7030 (Ø0.6955 ± 0.0075)
    At min CHF DIA, ang can be 119.500°-120.500°
    At nom CHF DIA, ang can be 124.458°-115-668°
    at max CHF DIA, ang can be 127.958°-119.500°
    Thanks in advance.
    Hi Dan,
    First,I would say that values are strange...( max < min ?) and it seems that nominal dia is missing.
    I would :
    ASSIGN/TOL_ARRAY=(119.5,120.5,115.668,124.458,119.5,127.958)
    ASSIGN/DELTA_D=(0.7030-0.6880)/3
    IF CHF_DIA>=0.6880 AND CHF_DIA<=(0.6880+DELTA_D)
    ASSIGN/TOL_MIN=TOL_ARRAY[1]
    ASSIGN/TOL_MAX=TOL_ARRAY[2]
    END IF
    IF CHF_DIA>(0.6880+DELTA_D) AND CHF_DIA<=(0.6880+2*DELTA_D)
    ASSIGN/TOL_MIN=TOL_ARRAY[3]
    ASSIGN/TOL_MAX=TOL_ARRAY[4]
    END IF
    IF CHF_DIA>(0.6880+2*DELTA_D) AND CHF_DIA<=(0.6880+3*DELTA_D)
    ASSIGN/TOL_MIN=TOL_ARRAY[5]
    ASSIGN/TOL_MAX=TOL_ARRAY[6]
    END IF
    IF CHF_DIA<0.6880 OR CHF_DIA>(0.6880+3*DELTA_D)
    COMMENT/OPER
    "Part is OOT !"
    END IF
    Then use tol_min and tol_max in tol values.
    • DAN_M
      DAN_M commented
      Hi Jeff, this is great, unfortunately not quite what I am looking for. You have this broken down by RANGE (which is about as far as I got). My engineers shot me down & would like it to come up with EXACT tolerance. So if the CHF was Ø0.7015, they'd like the EXACT calculated amount of allowable tolerance for the angle, not a "high limit range" for the angle. Any ideas? This is above any beyond my math skills to come up with a formula for this and then translate that into PC DMIS. =/
    Ok, I understand.
    The problem is that the tol is not really linear.
    Do they know it ?
    (I changed you values to
    0,688 119,5 120,5
    0,6955 124,458 125,668
    0,703 127,958 129,5
    Which seems to be more logical (maybe I'm wrong !)
    • DAN_M
      DAN_M commented
      I triple checked my numbers and they are correct. I know this is less logical and makes it more difficult but this is what I've got =/
    • JEFMAN
      JEFMAN commented
      DAN_M : I don't have any doubt about your checking, just about the customer values !!!!!
      How can the max be lower than the min ?
      How can you have a tol of 1° at the min, and close to 9° at the nom or the max dia ???????
      I would call the customer to its own check !!!!!
    • vpt.se
      vpt.se commented
      Gotta agree with JEFMAN , those numbers makes no sense.
    Let say you have 2 segments for each tolerance.
    You have for each one 2 points (min dia , max dia, min angle, max angle)
    You can write
    min angle=a* min dia + b
    max angle=a*max dia + b
    A and b are unknown, but can be calculated :
    a=(max angle - min angle)/(max dia - min dia)
    b= min angle- a * min dia.
    The tol angle for meas dia is a* meas dia + b.
    So with some assignments :
    ASSIGN/V1min=(119.5,124.458,,127.9 58)
    ASSIGN/V1max=(120.5,125.668,129.5)
    ASSIGN/V2=(0.688,0.6955,0.703)
    ASSIGN/A1MIN=(V1MIN[2]-V1MIN[1])/(V2[2]-V2[1])
    ASSIGN/B1MIN=V1MIN[1]-A1MIN*V2[1]
    ASSIGN/A1MAX=(V1MAX[2]-V1MAX[1])/(V2[2]-V2[1])
    ASSIGN/B1MAX=V1MAX[1]-A1MAX*V2[1]
    ASSIGN/A2MIN=(V1MIN[3]-V1MIN[2])/(V2[3]-V2[2])
    ASSIGN/B2MIN=V1MIN[2]-A2MIN*V2[2]
    ASSIGN/A2MAX=(V1MAX[3]-V1MAX[2])/(V2[3]-V2[2])
    ASSIGN/B2MAX=V1MAX[2]-A2MAX*V2[2]
    IF CHF_DIA>=0.6880 AND CHF_DIA<=0.6955
    ASSIGN/TOL_MIN=A1MIN*CHF_DIA+B1MIN
    ASSIGN/TOL_MAX=A1MAX*CHF_DIA+B1MAX
    END IF
    IF CHF_DIA>0.6955 AND CHF_DIA<=0.703
    ASSIGN/TOL_MIN=A2MIN*CHF_DIA+B2MIN
    ASSIGN/TOL_MAX=A2MAX*CHF_DIA+B2MAX
    END IF
    To be checked because I'm not sure of values, and here, it's time to go home ...
    Hope this helps.
      This may or may not be the most efficient way to do this. But it does work.
      CON1 =FEAT/CONTACT/CONE/DEFAULT,CARTESIAN,IN
      THEO/<0,0,0>,<0,0,1>,120,-0.1,0.6955
      ACTL/<0,0,0>,<0,0,1>,120,-0.1,0.6955
      TARG/<0,0,0>,<0,0,1>
      START ANG=0,END ANG=360
      ANGLE VEC=<1,0,0>
      SHOW FEATURE PARAMETERS=NO
      SHOW CONTACT PARAMETERS=NO
      DIM LOC1= LOCATION OF CONE CON1 UNITS=IN ,$
      GRAPH=OFF TEXT=OFF MULT=10.00 OUTPUT=NONE HALF ANGLE=NO
      AX NOMINAL MEAS +TOL -TOL DEV OUTTOL
      D 0.6880 0.6955 0.0150 0.0000 0.0075 0.0000 --------#---------
      END OF DIMENSION LOC1
      ASSIGN/V1=LOC1.D.DEV/.0075
      ASSIGN/V2=(LOC1.D.DEV/.0075)-1
      IF/V1<1
      ASSIGN/LOWEROUTPUT1="REPORT"
      ASSIGN/LOWERMIN=-((((115.668-119.50)*V1)+119.5)-120)
      ASSIGN/LOWERMAX=((((124.458-120.5)*V1)+120.5)-120)
      END_IF/
      IF/V1>1
      ASSIGN/LOWEROUTPUT1="NONE"
      END_IF/
      DIM LOC2= LOCATION OF CONE CON1 UNITS=IN ,$
      GRAPH=OFF TEXT=OFF MULT=10.00 OUTPUT=LOWEROUTPUT1 HALF ANGLE=NO
      A 120.0000 120.0000 LOWERMAX LOWERMIN 0.0000 0.0000 -#----------------
      END OF DIMENSION LOC2
      IF/V1>1
      ASSIGN/UPPEROUTPUT1="REPORT"
      ASSIGN/UPPERMIN=-((((119.5-115.688)*V2)+115.688)-120)
      ASSIGN/UPPERMAX=((((127.958-124.458)*V2)+124.458)-120)
      END_IF/
      IF/V1<1
      ASSIGN/UPPEROUTPUT1="NONE"
      END_IF/
      DIM LOC3= LOCATION OF CONE CON1 UNITS=IN ,$
      GRAPH=OFF TEXT=OFF MULT=10.00 OUTPUT=UPPEROUTPUT1 HALF ANGLE=NO
      A 120.0000 120.0000 UPPERMAX UPPERMIN 0.0000 0.0000 -#----------------
      END OF DIMENSION LOC3
        JEFMAN what do you think?
          You're right, your calculation gives the right tolerance, mine gives the boundaries !!!!!
          I didn't thought subtracting the nominal !!!!!!!