So this is what I find out after a gage R&R study. Every single feature was poor (high) for reproducibility and repeatability was very good (low). Obviously this is due to the loading and unloading of the part in the fixture, when the parts is not unloaded then repeatability is good but when is unloaded by a different operator then variation are present and minitab interpret this as “reproducibility”. And a little out of topic here but the variation btw is minimal but since is higher than repeatability and part-to-part variation then it appears to be very high and this is why gage r&r is not useful for me but nonetheless mandatory by customer.
What I did then was to redo the gage R&R with one single operator now loading and unloading after every run and of course the results were now very poor repeatability, meaning that it’s not really the difference between operator but the action of loading and unloading the part in the fixture which will never be perfectly placed in the same previous position.
Would you agree then that I should just run a repeatability? Does this makes more sense to you?