A previous employee snapped off one of our 1mm Probes and ended up changing some things in the calibration settings because we were not sure what the proper length of the probe was that he broke. Fast forward to now, we've replaced this probe with the previous length and I find that even after calibrating correctly, the probe was not taking points in the correct "Z" due to the length. The only solution I could think of was to measure the distance between the two probes we have and make the correction in the "Z" offset for the actual probe. It seems to have fixed our problem but is this the correct thing to do when you have a situation like that?
I am a 15 year Calypso programmer and when it comes to switching out probes even if it is a different length, all you have to due is re-qualify the stylus and it will generate the "X", "Y" & "Z" locations.
I am a 15 year Calypso programmer and when it comes to switching out probes even if it is a different length, all you have to due is re-qualify the stylus and it will generate the "X", "Y" & "Z" locations.