this is the error i am getting. what does it mean and how do i fix it?
invalid reading of the measured point
Collapse
X
-
It's normally when you've manually loaded a styli onto an analogue (scanning) probe but haven't performed a weight balance. Can also happen after a collision. To manually perform a weight balance, press the probe enable button twice (once to disable, once to enable again). An automatic weight balance will be performed whenever you index the wrist or do a tool change from a probe rack.
- Likes 1
-
Comment
-
What's your prehit/retract value and your touchspeed? For analogue probes the touchspeed should be 2mm/sec. You also need to make sure that you calibrated with that touchspeed - the touchspeed value within the probe utilities dialogue should always be the same as the value in your programs.
-
For prehit and retract, make sure the probe isn't hitting the surface too soon. If your prehit is set too small and the part has a large amount of deviation the tip may already be in contact with the surface when it enters it's probing cycle. Increase the prehit & retract if that is the case.
-
Related Topics
Collapse
-
by sscarpielloGreetings to all,
I am using for the first time a Leitz PMM 866 machine, yesterday the machine had an error trying to grab a different...-
Channel: PC-DMIS for CMMs
08-26-2011, 09:43 AM -
-
by johnrJust got a SCR200 rack hooked up and working fine. Only one question, when changing a probe the CMM goes straight to the load point from it's actual location...
-
Channel: PC-DMIS for CMMs
06-03-2014, 10:13 AM -
-
by TinyfxdsI have a 1.5mm x 20 mm probe tip that records a hit as soon as I press down on the "move" button after switching to slow mode. I am no where...
-
Channel: PC-DMIS for CMMs
07-24-2014, 02:54 PM -
-
by joshmxzI am using Renishaw probe changer and for a new part I need a 475+mm probe. The problem I have encountered a problem with loading the probe that I have...
-
Channel: PC-DMIS for CMMs
10-26-2012, 09:22 AM -
-
by Burr2222I'm having trouble setting up my MCR20 TP20 probe rack, it calibrates just fine, but when it goes to change probes it seems to mis-locate the probe in...
-
Channel: PC-DMIS for CMMs
06-19-2018, 10:44 AM -
Comment