invalid reading of the measured point

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • invalid reading of the measured point

    this is the error i am getting. what does it mean and how do i fix it?

  • #2
    It's normally when you've manually loaded a styli onto an analogue (scanning) probe but haven't performed a weight balance. Can also happen after a collision. To manually perform a weight balance, press the probe enable button twice (once to disable, once to enable again). An automatic weight balance will be performed whenever you index the wrist or do a tool change from a probe rack.
    Neil Challinor
    PC-DMIS Product Owner

    T: +44 870 446 2667 (Hexagon UK office)
    E: [email protected]

    Comment


    • Kp61dude!
      Kp61dude! commented
      Editing a comment
      So that's what it is called. Now I know.

  • #3
    I tried this and it didn't fix anything. how about any other tricks?

    Comment


    • neil.challinor
      neil.challinor commented
      Editing a comment
      What's your prehit/retract value and your touchspeed? For analogue probes the touchspeed should be 2mm/sec. You also need to make sure that you calibrated with that touchspeed - the touchspeed value within the probe utilities dialogue should always be the same as the value in your programs.

    • neil.challinor
      neil.challinor commented
      Editing a comment
      For prehit and retract, make sure the probe isn't hitting the surface too soon. If your prehit is set too small and the part has a large amount of deviation the tip may already be in contact with the surface when it enters it's probing cycle. Increase the prehit & retract if that is the case.

  • #4
    .06 for the pre/hit and retract and 2mm/sec for the touchspeed.

    Comment


    • neil.challinor
      neil.challinor commented
      Editing a comment
      Try increasing your prehit - as a test, set it to a large value like 0.25 if you can (I assume you're working in inches - if not then it's definitely too small - I'd make it at least a few mm)

  • #5
    it doesn't seem like anything is working.

    Comment


    • neil.challinor
      neil.challinor commented
      Editing a comment
      Sorry to hear that. I suggest contacting your local Hexagon support.

  • #6
    Is this a scanning probe or TTP? If scanning check you optional probe parameters and be sure to have an optional probe command before the 1st load probe command in the routine.

    Comment