I've got a weird one today: I calibrated my Master at A0B0 and several other probes. Then I get to my 1by35.5 Shank probe: A0B0 goes fine. A90 B0, A90 B90, A90 B180 all go fine. Then, A90 B-90 comes along and plunges straight down on the cal sphere about 3/5 of the way up the stem/shank. I have no ideas why this could be. The attached image has my settings for the calibration.
Probe Calibration Crash
Collapse
X
-
Have you tried 'reset tips to theo' check box?
For many different reasons, probes have seemingly lost/shifted information. Often that check box will clear the issue up.
-
What exactly does Reset Theo do? Is it just for the angles selected that it works? Would me using it everytime be an issue?
-
When you first define a probe, the software uses the Theoretical probe offsets to drive to and around the calibration sphere. Once a probe is calibrated, it has a more accurate definition of the offsets, and uses these new values to re-calibrate the sphere each time after that. The "Reset Theos..." tells it NOT to use the prior calculated offsets but instead to use the probe definition's nominal offsets.
-
-
I've had this happen before. Turns out it was a corrupted probe file. Shut down DMIS, go into your probe files folder and delete your 1by35.5 probe file. Restart DMIS and the measurement routine, F9 your loadprobe line for the 1by35.5 probe and rebuild it then do the calibration procedure again. Had fixed it every time for me.Remembering my beautiful wife Taz who's life was lost on 6-13-2020. I love you and I miss you.
Comment
Related Topics
Collapse
-
by Ken8282I came in after the weekend and went to calibrate the probe and it kept contacting the sphere and flexing the probe to much so I got an unexpected probe...
-
Channel: PC-DMIS for CMMs
11-05-2018, 02:26 PM -
-
by tnguyenHi.
It had something changes calibration by Operator yesterday. Today, I calibrate Master Probe A0B0, It's runing fine, then calibrate all probes,...-
Channel: PC-DMIS for CMMs
05-08-2011, 04:50 PM -
-
I have a issue where I have 6 different probes set up in a tool changer, however there is no calibration program for this machine, the calibration method...
-
Channel: PC-DMIS for CMMs
07-31-2012, 09:09 AM -
-
by cmhallHi, So when calibrating probe it is like I'm 10mm off.I calibrate my probe at A0,B0 ok but then it moves to A90,B-90 its hitting the calibration sphere.So...
-
Channel: PC-DMIS for CMMs
01-13-2023, 05:06 AM -
-
by angerjiGood Afternoon,
Working on a program as we speak and have what I hope to be a pretty simple concept and question, but having a hard time...-
Channel: PC-DMIS for CMMs
10-29-2018, 03:50 PM -
Comment