6 Way Probe Mod
  • 6 Way Probe Mod
    Hi all,
    Couple of questions regrading the 6way star mods...
    1) I use TesastarMp_US modules. Would I be able to use a 6way TP20 on my machine? I am not seeing a 6way mod online that is specifically for Tesastar.
    2) What sort of accuracy are you seeing on these? (I am really hoping to be able to hold ±0.002 with this mod. Ideally I'd like to make a 20mm ext, a 5 way star, then a 1X10 sticking out of tip 3)
    how does a 6way work?
    • DAN_M
      DAN_M commented
      idk check it out
      The TP20 is a compact module changing touch-trigger probe. The two-piece design, comprising of a probe body and detachable stylus module(s), allows the use of a range of stylus configurations and extensions to access features on complex parts. Modules can be changed manually or automatically without the need for requalification of the stylus tip.The 6-way probe module is identified by a blue cap. This probe module is designed for 6-way operation where there is a requirement to measure in the –Z direction, for example when measuring undercuts.The recommended stylus limits for the 6-way probe module are:•Any stylus type up to 30 mm long•Star and cranked styli up to 10 mm offset
    6 way is for a complete star probe. You can measure x-y-z- + and - from all 5 tips / directions. ( 1 probe at all the 90's and 1 facing straight down like standard probe )
      Yes, Hexagon TesaStar's are interchangeable (rebranded renishaw) with TP20
      Renisaw TP20 data sheet: https://www.renishaw.com/en/tp20--6670
      Says the 6W is about 2x worse than standard force in repeatability (0.00004") and only 0.00004" worse in pretravel variation.
      I have an old spec sheet, with other data:
      TP20 probe technical data
      NOTE: For further information on the TP20 probe, refer to the TP20 system installation and user’s guide (Renishaw
      part number H-1000-5008), available through your supplier.
      NOTE: The following data is derived from high accuracy test rig measurements and may not represent the
      performance achievable on a CMM. Please consult you CMM supplier for overall system accuracy information.
      Performance at 10 mm stylus length
      (You'll have to CSV this or something to read it better, it's in a table that i can't paste. I BOLDED the 6W data)
      Parameter
      Probe module type
      LF SF MF EF 6W EM1 STD EM2 STD
      Unidirectional repeatability* (2σ) 0.35 μm 0.35 μm 0.50 μm 0.65 μm 0.8 μm 0.35 μm 0.35 μm
      2D (XY) form measurement deviation* ±0.6 μm ±0.8 μm ±1.00 μm ±2.00 μm ±1.5 μm ±0.8 μm ±0.8 μm
      Measured at trigger speed of 8 mm/s.
      Test stylus ball Ø4 mm.
      Without extensively testing it, I presume your uncertainty of that probe module will be about 0.0001" worse than the uncertainty of a standard force module.
        We use a 6way. StdDev on my normal A0B0 probes are about 0.001mm whereas my star probe is anywhere from 0.002mm to 0.003mm.
          Other star probes and builds can’t check under cuts or take hits in Z- as good unless it’s a 6 way?
          • louisd
            louisd commented
            That's correct. TP20's do not have the ability to open the circuit (detect a point) in Z-. If you are probing with a TP20 w/star probe on any other module, you are detecting the hit when the probe travels enough to open the circuit in X or Y, on the opposing side of the module from where you are taking the hit.
          • acgarcia
            acgarcia commented
            Thanks for the clarification. i'm going to have to think if I've done this before. I tend to use extended force modules so its a bummer that the max length of the 6 way is 30mm.
          We run the TP20's exclusively and we use the 6 way mods for angle probes. Typical STD Deviation is .0001"-.0003" all day long.
            louisd
            I just got a 6 way probe. Set it all up. My machine won't get it from my rack =/ I have no idea why. Are you positive that I can use this?
            My usual setup: HH-A-T5, HA-TM-31, TesastarMP_Body_US
            What I built: HH-A-T5, HA-TM-31, Probe TP20_6Way, 20mm ext, 5way test star, 1X10 in tip 3
            I my new probe in the rack, told the rack it was in a port, when I tell PC DMIS to get the tool it wants me to hand load the probe
            • louisd
              louisd commented
              Did you assign the probe to the rack position?
              If so, and it didn't save...
              I came to the same roadblock when I tried to use a TP20 MCR20 rack to hold TP200 modules. It's Hexagon/Renishaw's ploy to mandate you to keep in-family components. Even though the TP200 physically docks/undocks with success on the MCR20, they essentially utilized this software-lock to force us to get PI200 racks.
              Create the 6-way probe as your normal TesastarMP_Body_US, and you should be fine. Unlike the TP20/TP200, your physical dimensions between the TesastarMP and the TP20 should be identical. The difference is internal to the module. There is nothing different between a TP20 SF and a TP20 6way, in regards to how the software considers the probe, but there might be a sw lock on using a TP20 on whatever branding the Tesastar Rack is.