    Anyone know how to solve this problem? “Waiting for response : GETPOS”
    i always getting this pop-up when creating new program. I have to close the pcdmis then restart the machine to settle the error. Anyone know why this happen?

    software : 2019R1
    controller : FB2
    probe head : PH10M
    force sensor : TP200

  • #2
    I would try to home the cmm.
    (edit/cmm setup/terminal/type "autzer").
    I think that the cmm loose the count (I don't know why...), so homing helps to re-find it.

    Comment


    • Shahidan Firdaus
      Shahidan Firdaus commented
      Editing a comment
      But after certain time it happen again

  • #3
    Try to clean the scales, check the wiring...

    Comment


    • Shahidan Firdaus
      Shahidan Firdaus commented
      Editing a comment
      Done with this but still have the problem

  • #4
    Or check the Energy Savings set in Windows - don't let Windows turn off (or lower) power to USB ports, serial ports or network cards, no matter how long the computer has been 'unused' (Windows often seems to only consider keyboard and mouse as 'use').
    AndersI
    SW support - Hexagon Metrology Nordic AB

    Comment


    • #5
      Also disable FIFO buffering on the serial card.
      Systems Integrator
      Hexagon Manufacturing Intelligence

      Comment


      • Shahidan Firdaus
        Shahidan Firdaus commented
        Editing a comment
        Disable FIFO also cannot solve

    • #6
      Did you try to type "status" in the terminal ?
      What is the answer ?

      Comment


      • Shahidan Firdaus
        Shahidan Firdaus commented
        Editing a comment
        I just solve the problem by clicking “use TCIP” and set the IP address. Thanks for your support