Hello Everyone!
I was searching the forum but could not find solution for this specific problem:
We have no clue what to do. We predict that Renishaw will just say that it works for them.
If anyone knows any tips or solutions it would be great and I would be forever grateful.
IMG_20191022_105044.jpgIMG_20191022_105056.jpg
I was searching the forum but could not find solution for this specific problem:
- I’m using DEA GLOBAL Classic with HH-MI-M Manual Probe Head and Renishaw TP20 module.
- This specific TP20 module is used on 2 different machines (2x DEA GLOBAL classic). It is calibrated with both machines
- Today this TP20 stopped working on one of the machines. There is no connection, LED light won’t lit and probe won’t register hits
- We thought that it is HH-MI-M problem but other modules we use on this machine are working fine, so that excludes Manual Probe Head as the culprit.
- This TP20 works on the other machine with no problem.
- Crash Pins on TP20 are hidden
- When we squeeze the Manual Probe Head to change the angle, the light turns on and the TP20 registers hits with no problem, but when we let go of the Head it disconnects (I attached photos to show the problem)
- If it wouldn’t work on both machines the cause would be the defective TP20. If the Manual Probe Head would malfunction then all the modules wouldn’t work with it. We are at loss.
- TP20 is not measuring magnetized parts or parts fixed with magnets
- Demagnetization of the TP20 did not help
We have no clue what to do. We predict that Renishaw will just say that it works for them.
If anyone knows any tips or solutions it would be great and I would be forever grateful.