  • FCR25 Probe Rack Configuration Changes

    I need to configure an FCR25 probe rack to drop off the SM25-1, SM25-2 modules into each slot, rather than dropping of the SH25-1, SH25-2 modules, as it currently does. Has anyone had to make this change before and could offer some guidance?TIA
    3.7mr3CAD++ / 2011mr1CAD++/2012mr1CAD++/QUINDOS7

  • #2
    Not sure what you are looking for, but

    a) you need to remove any inserts
    b) you need to re-define and re-calibrate the rack

    Its really not too difficult. Have you tried it and run into some specific question?

    Comment


    • #3
      When you physically remove the inserts from the slot also edit the slot in PC-DMIS to be No Insert.

      As a side note, it will only drop off the SM25 module when it needs to change to a different SM25 module.
      PC-DMIS 2016.0 SP8

      Jeff

      Comment


      • #4
        Originally posted by Schrocknroll View Post
        When you physically remove the inserts from the slot also edit the slot in PC-DMIS to be No Insert.

        As a side note, it will only drop off the SM25 module when it needs to change to a different SM25 module.
        Thanks for the reply. So are you saying that if I physically have say 2 SM25-1 modules in the rack that it won’t change between both SM25-1 modules? TIA
        3.7mr3CAD++ / 2011mr1CAD++/2012mr1CAD++/QUINDOS7

        Comment


        • #5
          Originally posted by Liambo734 View Post

          Thanks for the reply. So are you saying that if I physically have say 2 SM25-1 modules in the rack that it won’t change between both SM25-1 modules? TIA
          You can switch between 2 SM25-1, if you want to. You assign which probes will be used with which module. If you assign one probe to SM25-1 in slot #1 and assign a second probe to SM25-1 in slot #2, it will switch between the 2 SM25-1 modules when changing the two probes.

          The key to remember is this, each probe build will need to be assigned to 2 slots, the one with the module to be used and the one the probe will be stored in.

          Let's say you have a SM25-1 in slot #1 and a SM25-1 in slot #2. Then you have probe builds #1&2 using a SH-1. If you want probe build #1 in slot #3 using SM25-1 #1 and probe build #2 in slot#4 using SM25-1 #2, this is what you do. Physically put a SM25-1 in slots 1&2. Physically put build #1 in slot 3 and build #2 in slot 4. Then in PC-DMIS assign build #1 to slots 1 and 3. Assign build #2 to slots 2 and 4. Then PC-DMIS will know that for build #1 it will use the module in slot 1 and the probe in slot 3. For build #2 it will use the module in slot 2 and the probe in slot 4. It doesn't care if it's a -1, -2, or -3. It just knows that since there isn't an insert in 1 & 2, they are modules and since there is an insert in sots 3 & 4 they are probe builds(once again, this needs to be defined in PC-DMIS and physically done in the slots).

          I hope that wasn't too wordy, but I hope that helps.
          PC-DMIS 2016.0 SP8

          Jeff

          Comment