Again I am new for disk probes is there a preferred angle range being they are not spherical it puzzles me.
Calibration Start and End Angles for disk probes
Collapse
X
-
hello
you can use this calculation to figure out ur start-end angle.
example:
18mm disc dia x 2mm thick
Calibrate using only 70% of the disk thickness (out of square comp.)
2mm x .70 = 1.4mm
half of 70% thickness
1.4mm / 2 = .7mm
Half probe dia.
18mm/2=9mm
Calculate the angle from horizontal center line to half disk thickness
SineA= .7/9
A=4.4º
Start angle -4.4º
End angle 4.4º
- Likes 2
Comment