You just have to insert a "loadprobe" command...
If you use more than 1 probe (tip or probe ?), you have to define a "master probe" and calibrate all the tips / probes on the same sphere.
You can look on the forum about master probe and calibration...
You can also look at the top of the forum, in the support / dowload section, there are some very intersting informations.
Are you changing probes manually, or is there a programmed cycle that will change out your probes from a rack?
If it's manual, you just have to let the CMM know (LoadProbe) that a change has taken place. We've got a unit here that doesn't have a changer, and I usually just insert a ReportCommand/Comment to let the operator know when to put the new probe on the unit. They hit the enter key when they've done it, and the program continues...
If it's got a probe change rack, it's a matter of doing a LoadProbe, and it should take care of it for you.
In either case, you've programmed the difference in probe parameters. Easy Peasy, lemon squeasy
WIth a probe rack, you need to define your probes (LOADPROBE, then press F9. Build probe to match), then calibrate them.
Once calibrated you can go to Edit -> Preferences -> Probe Changer. Once there, hit F1 for more guidance.
You will need to:
calibrate your rack (probably an MCR20), -- any probe works for the cal, but Hex recommends using a 2x20 probe for the MCR20 rack cal
determine a mount point (which is basically a clearance movepoint, that the CMM moves to, before and after changing probes),
then assign the probes to each rack position.
Once the probes are assigned, cycle through all the probes via loadprobe commands, and manually place each probe onto the head. Let the CMM 'park' each probe on its own. After this, you are set on having a functional auto-changing rack.
I have a issue where I have 6 different probes set up in a tool changer, however there is no calibration program for this machine, the calibration method...
Hi. I'd like to be able to calibrate the 4 different touch probes (for all the existing rotations) that are stored in my probe rack using a single program...
Comment