Calibration of probe diameter way off
  • Calibration of probe diameter way off
    TRIED THIS WITH VOLCOMP METHOD 14 AND 0
    SHARP 2 CONTROLLER
    CAL TOOL IS 19.05... CHECKED IT
    TRIED IT WITH 2X20 THEN 5X30
    THESE ARE THE 5X30 RESULTS
    Probe file=5X30 Date=5/26/2019 Time=4:13:35 PM
    NEW SPHERE CENT X 1093.193 Y 2850.023 Z -773.128 D 19.050
    T1A0B0 THEO X 0.000 Y 0.000 Z 171.520 D 4.000
    T1A0B0 MEAS X 0.000 Y 0.000 Z 171.520 D 96.066 StdDev 0.008
    Measured probe diameter error exceeds the limits.
    T1A0B0
    LIL HELP
    2019 R1
    Correct probe, loose probe, dirty probe, bad module. Once had a tip where the glue wasn't sticking to the ruby. Just enough glue around the hole for the stem so that the ball wouldn't fall off.
      I would check the unit (metric or imperial) of the creation of the sphere and the tip (4*25.4 not too far of 96...)
        changed tp20 body and module
        started new program in mm
        changed location of probe directory
        rebuilt probe ph10mq , convert 30mm to 8mm for autojoint (old) tp20 probe 1x10
        rebuilt cal tool 0,0,1,19.05... confirmed size
        results
        Probe file=1X10 Date=5/27/2019 Time=6:00:24 AM
        newone CENT X 1094.990 Y 2848.865 Z -852.951 D 19.050
        T1A0B0 THEO X 0.000 Y 0.000 Z 151.520 D 1.000
        T1A0B0 MEAS X 0.000 Y 0.000 Z 151.520 D 81.119 StdDev 0.001
        T1A0B0
        I don't see where to change the units except when creating the program?
          What CMM are you using ?????
          The sphere location at Y 2850 mm ?
          Reset the tip, answer yes to "did the sphere moved", and tell us some news...
          In addition, check if you have write rights in the folder of tool.dat.
            cmm is a microxcel 7-10-5 pfx
            I moved the cal sphere to 3228 and the data showed that
            reset the tip
            NOT SURE ABOUT WRITE RIGHTS BUT TOOL .DAT IS IN THE INSTALL FOLDER F:\Program Files\PC-DMIS 2019 R2 64-BIT 7105
            Probe file=1X10 Date=5/27/2019 Time=7:17:55 AM
            new ball CENT X 1094.828 Y 3228.029 Z -852.833 D 19.050
            T1A0B0 MEAS X 0.000 Y 0.000 Z 151.520 D 81.114 StdDev 0.000
            T1A0B0
              The only thing that is out is the probe diameter??? Never seen this
              BTW RUNNING 2019 R1
                reset the tip
                How can you move the sphere to Y=3228 on a cmm which has a Y axis of 1000 ?????
                I would rename the tool.dat of the install folder as tool.old, then I would try to re-create the sphere at all....
                C:\ProgramData\HEXAGON\PC-DMIS\2019 R1 : tool.dat should be here...
                  As I just purchased this machine used and was given a usb with what I think is bum Downl>oad and Comp.dat files ... I thank ou for your efforts adn will get back to you once I get the correct files
                  Thanx for all your help so far Jeff
                    You're welcome, I hope you will find the issue...