Deviation with off-line programming?
  • Deviation with off-line programming?
    I’m programming offline to a model. Occasionally I see deviation in the CMM report of a feature. When I go back and look at the points, I sometimes see a difference between THEO’s and ACTL’s.
    I right click them and pick “Reset measured values to nominal” refresh the feature and refresh the dimensions and they read correct.
    What’s causing the deviation?
    Thanks
    CAD++ Version 2017 R1
    Build 1313 SP7
    CAD++ 2017 R1, Catia Translator
    B&S Xcel, Renishaw PH10MQ
    Without seeing your code, check to see that your alignments lock all 6 degrees of freedom. That can cause your nominals to get messed up
      I’m doing iterative alignments. Generally, level to 4 points, rotate to 2, origin 1 point. My nominals aren’t messed up; my actuals shown deviation.
        well, iterative alignments re-solve when you re-open a program, even when you do not run them. I would guess this is the issue. all the rounding issues when going from 12-place cad data to 3 (or 4 or 5) place program values.
          It might have to do with your iterative..
          You only need THREE hits to level, two to rotate, and one to origin. Adding additional "leveling" hits induces "fixture error" and could be messing with your nominals.
          Something alignment related is going on here...if your alignments are correct & so is your CAD model then everything should always be zero/nominal when you run offline.. :-/
          • measurestuff413
            measurestuff413 commented
            Can you post the code??
          Maybe...if you tell me how.
          Copy/Paste gives an error it's too many characters.
            Multiple posts if your code can't fit is ok,
            [code] put your code here [ /code]
            --> don't put a space between the [ /code]
              I can’t show the CAD. But if you look at the attached (maybe) PDF report it shows SF2 with .001 deviation. Surface 2 is the B datum. It is a plane constructed out of PNT5,PNT6,PNT9. PNT5 &6 are my 2 points of rotation for my alignment. PNT 9 is the 3rd point in a plane constructed after the alignment. PNT9 shows deviation.
              SF1 is the A datum. It’s a non-planer surface that I have constructed a plane out of with PNT2,3 &4. I also level to those 3 points. SF2 and SF3 (B&C) are planer surfaces constructed with 3 points.
              Surfaces 15, 16 & 17 all showed .001 deviation. None of the hits show deviation, so I switched to 4 places. Still no deviation on the points.
              This example doesn’t have much deviation, but in some of these I see as much as .006. Until I reset the actuals to nominal.
              I'll try to post it next....
              EDIT: Okay, I had to cut a whole bunch of stuff out to get this to post.... but the problem areas are there.
                startup =alignment/start,recall:use_part_setup,list=yes
                alignment/end
                mode/manual
                touchspeed/ 0.75
                format/text,options, ,headings,symbols, ;nom,tol,meas,dev,outtol, ,
                loadprobe/2_50
                tip/t1a0b0, shankijk=0, 0, 1, angle=0
                displayprecision/4
                man_pnt1 =feat/point,cartesian
                theo/<0.15,4.4,0.006>,<-0.0463915,-0.0062763,0.9989036>
                actl/<0.15,4.4,0.006>,<-0.0463915,-0.0062763,0.9989036>
                meas/point,1,workplane
                hit/basic,normal,<0.15,4.4,0.006>,<-0.0463915,-0.0062763,0.9989036>,<0.15,4.4,0.006>,use theo=yes
                endmeas/
                man_pnt2 =feat/point,cartesian
                theo/<3.2,0.8,-0.1938>,<-0.0000593,-0.0145345,0.9998944>
                actl/<3.2,0.8,-0.1938>,<-0.0000593,-0.0145345,0.9998944>
                hit/basic,normal,<3.2,0.8,-0.1938>,<-0.0000593,-0.0145345,0.9998944>,<3.2,0.8,-0.1938>,use theo=yes
                endmeas/
                man_pnt3 =feat/point,cartesian
                theo/<0.15,-4.4,0.0006>,<-0.008504,0.0036002,0.9999574>
                actl/<0.15,-4.4,0.0006>,<-0.008504,0.0036002,0.9999574>
                hit/basic,normal,<0.15,-4.4,0.0006>,<-0.008504,0.0036002,0.9999574>,<0.15,-4.4,0.0006>,use theo=yes
                endmeas/
                man_pnt4 =feat/point,cartesian
                theo/<-0.0101,-4.4,-0.07>,<-0.9896744,0,0.1433341>
                actl/<-0.0101,-4.4,-0.07>,<-0.9896744,0,0.1433341>
                hit/basic,normal,<-0.0101,-4.4,-0.07>,<-0.9896744,0,0.1433341>,<-0.0101,-4.4,-0.07>,use theo=yes
                endmeas/
                man_pnt5 =feat/point,cartesian
                theo/<-0.01,4.4,-0.07>,<-0.9896744,0,0.1433341>
                actl/<-0.01,4.4,-0.07>,<-0.9896744,0,0.1433341>
                hit/basic,normal,<-0.01,4.4,-0.07>,<-0.9896744,0,0.1433341>,<-0.01,4.4,-0.07>,use theo=yes
                endmeas/
                man_pnt6 =feat/point,cartesian
                theo/<0.4,4.5809,-0.07>,<0.0177283,0.9998377,-0.003204>
                actl/<0.4,4.5809,-0.07>,<0.0177283,0.9998377,-0.003204>
                hit/basic,normal,<0.4,4.5809,-0.07>,<0.0177283,0.9998377,-0.003204>,<0.4,4.5809,-0.07>,use theo=yes
                endmeas/
                a1 =alignment/start,recall:startup,list=yes
                alignment/iterate
                pnt target rad=0.25,start label=,fixture tol=0,error label=
                meas all feat=once,level axis=zaxis,rotate axis=xaxis,origin axis=yaxis
                level=man_pnt1,man_pnt2,man_pnt3,,
                rotate=man_pnt4,man_pnt5,,
                origin=man_pnt6,,
                alignment/end
                mode/dcc
                clearp/zplus,2.5,zplus,0,off
                move/clearplane
                pnt1 =feat/contact/vector point/default,cartesian
                targ/<0.4,4.5809,-0.07>,<0.0177283,0.9998377,-0.003204>
                snap=no
                show feature parameters=no
                show contact parameters=yes
                avoidance move=no,distance=0.3937
                show hits=yes
                hit/basic,<0.4,4.5809,-0.07>,<0.0177283,0.9998377,-0.003204>,<0.4,4.5809,-0.07>
                endmeas/
                move/clearplane
                tip/t1a97.5b90, shankijk=-0.9914, 0, -0.1305, angle=90
                move/point,normal,<-0.25,3,4>
                move/point,normal,<-0.25,3,-0.5>
                pnt2 =feat/contact/vector point/default,cartesian
                theo/<0.4,3,-0.2415>,<0.0383396,0.004845,-0.999253>
                actl/<0.4,3,-0.2415>,<0.0383396,0.004845,-0.999253>
                targ/<0.4,3,-0.2415>,<0.0383396,0.004845,-0.999253>
                snap=no
                avoidance move=both,distance=0.25
                show hits=yes
                hit/basic,<0.4,3,-0.2415>,<0.0383396,0.004845,-0.999253>,<0.4,3,-0.2415>
                endmeas/
                pnt3 =feat/contact/vector point/default,cartesian
                theo/<0.4,-3,-0.2503>,<0.012293,-0.0012943,-0.9999236>
                actl/<0.4,-3,-0.2503>,<0.012293,-0.0012943,-0.9999236>
                targ/<0.4,-3,-0.2503>,<0.012293,-0.0012943,-0.9999236>
                snap=no
                show hits=yes
                hit/basic,<0.4,-3,-0.2503>,<0.012293,-0.0012943,-0.9999236>,<0.4,-3,-0.2503>
                endmeas/
                pnt4 =feat/contact/vector point/default,cartesian
                theo/<3,-0.3,-0.3085>,<-0.0026184,0.0126152,-0.999917>
                actl/<3,-0.3,-0.3085>,<-0.0026184,0.0126152,-0.999917>
                targ/<3,-0.3,-0.3085>,<-0.0026184,0.0126152,-0.999917>
                snap=no
                show hits=yes
                hit/basic,<3,-0.3,-0.3085>,<-0.0026184,0.0126152,-0.999917>,<3,-0.3,-0.3085>
                endmeas/
                move/point,normal,<-0.25,-0.3,-0.5>
                move/point,normal,<-0.25,-0.3,7>
                move/clearplane
                pnt5 =feat/contact/vector point/default,cartesian
                targ/<-0.0101,-4.4,-0.07>,<-0.9896744,0,0.1433341>
                snap=no
                show hits=yes
                hit/basic,<-0.0101,-4.4,-0.07>,<-0.9896744,0,0.1433341>,<-0.0101,-4.4,-0.07>
                endmeas/
                move/clearplane
                pnt6 =feat/contact/vector point/default,cartesian
                theo/<-0.0101,4.4,-0.07>,<-0.9896744,0,0.1433341>
                actl/<-0.0101,4.4,-0.07>,<-0.9896744,0,0.1433341>
                targ/<-0.0101,4.4,-0.07>,<-0.9896744,0,0.1433341>
                snap=no
                show hits=yes
                hit/basic,<-0.0101,4.4,-0.07>,<-0.9896744,0,0.1433341>,<-0.0101,4.4,-0.07>
                endmeas/
                move/clearplane
                a2 =alignment/start,recall:a1,list=yes
                alignment/iterate
                pnt target rad=0.01,start label=,fixture tol=0,error label=
                meas all feat=always,max iterations=3,level axis=zaxis,rotate axis=xaxis,origin axis=yaxis
                level=pnt2,pnt3,pnt4,,
                rotate=pnt5,pnt6,,
                origin=pnt1,,
                alignment/end
                  PNT7       =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                              THEO/<0.7245,4.575,-0.1328>,<0.0177283,0.9998377,-0.003204>
                              ACTL/<0.7245,4.575,-0.1328>,<0.0177283,0.9998377,-0.003204>
                              TARG/<0.7245,4.575,-0.1328>,<0.0177283,0.9998377,-0.003204>
                              SNAP=NO
                              SHOW FEATURE PARAMETERS=NO
                              SHOW CONTACT PARAMETERS=YES
                                AVOIDANCE MOVE=BOTH,DISTANCE=0.25
                              SHOW HITS=YES
                                HIT/BASIC,<0.7245,4.575,-0.1328>,<0.0177283,0.9998377,-0.003204>,<0.7245,4.575,-0.1328>
                              ENDMEAS/
                  PNT8       =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                              THEO/<0.1179,4.5856,-0.1756>,<0.0177283,0.9998377,-0.003204>
                              ACTL/<0.1179,4.5856,-0.1756>,<0.0177283,0.9998377,-0.003204>
                              TARG/<0.1179,4.5856,-0.1756>,<0.0177283,0.9998377,-0.003204>
                              SNAP=NO
                              SHOW HITS=YES
                                HIT/BASIC,<0.1179,4.5856,-0.1756>,<0.0177283,0.9998377,-0.003204>,<0.1179,4.5856,-0.1756>
                              ENDMEAS/
                              MOVE/CLEARPLANE
                  PNT9       =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                              THEO/<-0.0065,-1.3542,-0.045>,<-0.9896744,0,0.1433341>
                              ACTL/<-0.0056,-1.3542,-0.0454>,<-0.9896744,0,0.1433341>
                              TARG/<-0.0056,-1.3542,-0.0454>,<-0.9896744,0,0.1433341>
                              SNAP=NO
                              SHOW HITS=YES
                                HIT/BASIC,<-0.0056,-1.3542,-0.0454>,<-0.9896744,0,0.1433341>,<-0.0056,-1.3542,-0.0454>
                              ENDMEAS/
                  SF1        =FEAT/PLANE,CARTESIAN,TRIANGLE,NO
                              THEO/<1.2667,-0.1,-0.2667>,<-0.0239176,0.0014707,-0.9997129>
                              ACTL/<1.2666,-0.0995,-0.2667>,<-0.0239151,0.0014705,-0.9997129>
                              CONSTR/PLANE,BF,PNT2,PNT3,PNT4,,
                              OUTLIER_REMOVAL/OFF,3
                              FILTER/OFF,WAVELENGTH=0
                  SF2        =FEAT/PLANE,CARTESIAN,TRIANGLE,NO
                              THEO/<-0.0089,-0.4514,-0.0617>,<-0.9896744,0,0.1433341>
                              ACTL/<-0.0087,-0.4511,-0.0618>,<-0.9832431,-0.0000085,0.1822992>
                              CONSTR/PLANE,BF,PNT5,PNT6,PNT9,,
                  SF3        =FEAT/PLANE,CARTESIAN,TRIANGLE,NO
                              THEO/<0.4141,4.5805,-0.1261>,<0.0177283,0.9998377,-0.003204>
                              ACTL/<0.4141,4.5806,-0.1261>,<0.0180955,0.9998008,-0.0084178>
                              CONSTR/PLANE,BF,PNT1,PNT7,PNT8,,
                              DATDEF/FEATURE=SF1,A
                              DATDEF/FEATURE=SF2,B
                              DATDEF/FEATURE=SF3,C
                              DISPLAYPRECISION/3
                  D1 =SURFACE PROFILE : SF1
                              FEATCTRLFRAME/SHOWPARAMS=YES,SHOWEXPANDED=YES
                              CADGRAPH=OFF,REPORTGRAPH=OFF,TEXT=OFF,MULT=10.00,ARROWDENSITY=100,OUTPUT=BOTH,UNITS=IN
                              COMPOSITE=NO,
                              VECTOR_MIN_MAX,ITERATEANDREPIERCECAD=NO
                              STANDARDTYPE=ASME_Y14_5
                                PRIMARY DIMENSION/PROFILE OF SURFACE,0.02,<UZ>,<UTol>,<dat>,<dat>,<dat>
                                NOTE/D1
                              FEATURES/SF1,,
                  D2 =SURFACE PROFILE : SF2
                              COMPOSITE=NO,
                              CUSTOMIZED DRF=NO
                                PRIMARY DIMENSION/PROFILE OF SURFACE,0.02,<UZ>,<UTol>,A,<dat>,<dat>
                                NOTE/D2
                              FEATURES/SF2,,
                  D3 =SURFACE PROFILE : SF3
                              COMPOSITE=NO,
                              CUSTOMIZED DRF=NO
                                PRIMARY DIMENSION/PROFILE OF SURFACE,0.02,<UZ>,<UTol>,A,B,<dat>
                                NOTE/D3
                              FEATURES/SF3,,
                    DISPLAYPRECISION/4
                    PNT43 =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                    THEO/<1.6724,2.8832,-0.0753>,<0.4256742,0.9048427,0.0078192>
                    ACTL/<1.6724,2.8832,-0.0753>,<0.4256742,0.9048427,0.0078192>
                    TARG/<1.6724,2.8832,-0.0753>,<0.4256742,0.9048427,0.0078192>
                    SNAP=NO
                    SHOW HITS=YES
                    HIT/BASIC,<1.6724,2.8832,-0.0753>,<0.4256742,0.9048427,0.0078192>,<1.6724,2. 8832,-0.0753>
                    ENDMEAS/
                    PNT44 =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                    THEO/<1.3027,3.1821,-0.0849>,<0.7954317,0.6059829,0.0085541>
                    ACTL/<1.3027,3.1821,-0.0849>,<0.7954317,0.6059829,0.0085541>
                    TARG/<1.3027,3.1821,-0.0849>,<0.7954317,0.6059829,0.0085541>
                    SNAP=NO
                    SHOW HITS=YES
                    HIT/BASIC,<1.3027,3.1821,-0.0849>,<0.7954317,0.6059829,0.0085541>,<1.3027,3. 1821,-0.0849>
                    ENDMEAS/
                    PNT45 =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                    THEO/<1.109,3.6407,-0.0762>,<0.9890567,0.1473574,0.007252>
                    ACTL/<1.109,3.6407,-0.0762>,<0.9890567,0.1473574,0.007252>
                    TARG/<1.109,3.6407,-0.0762>,<0.9890567,0.1473574,0.007252>
                    SNAP=NO
                    SHOW HITS=YES
                    HIT/BASIC,<1.109,3.6407,-0.0762>,<0.9890567,0.1473574,0.007252>,<1.109,3.64 07,-0.0762>
                    ENDMEAS/
                    SF15 =FEAT/SET,CARTESIAN
                    THEO/<1.3614,3.2353,-0.0788>,<0,0,1>
                    ACTL/<1.3614,3.2353,-0.0788>,<0,0,1>
                    CONSTR/SET,BASIC,PNT43,PNT44,PNT45,,
                    D15 =SURFACE PROFILE : SF15
                    CADGRAPH=OFF,REPORTGRAPH=OFF,TEXT=OFF,MULT=10.00,A RROWDENSITY=100,OUTPUT=BOTH,UNITS=IN
                    COMPOSITE=NO,
                    CUSTOMIZED DRF=NO
                    PRIMARY DIMENSION/PROFILE OF SURFACE,0.06,<UZ>,<UTol>,A,B,C
                    NOTE/D15
                    FEATURES/SF15,,
                    PNT46 =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                    THEO/<1.0986,3.8919,-0.1322>,<0.9999785,-0.0008633,0.0064954>
                    ACTL/<1.0986,3.8919,-0.1322>,<0.9999785,-0.0008633,0.0064954>
                    TARG/<1.0986,3.8919,-0.1322>,<0.9999785,-0.0008633,0.0064954>
                    SNAP=NO
                    SHOW HITS=YES
                    HIT/BASIC,<1.0986,3.8919,-0.1322>,<0.9999785,-0.0008633,0.0064954>,<1.0986,3.8919,-0.1322>
                    ENDMEAS/
                    PNT47 =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                    THEO/<1.0981,4.0521,-0.0354>,<0.9999785,-0.0008633,0.0064954>
                    ACTL/<1.0981,4.0521,-0.0354>,<0.9999785,-0.0008633,0.0064954>
                    TARG/<1.0981,4.0521,-0.0354>,<0.9999785,-0.0008633,0.0064954>
                    SNAP=NO
                    SHOW HITS=YES
                    HIT/BASIC,<1.0981,4.0521,-0.0354>,<0.9999785,-0.0008633,0.0064954>,<1.0981,4.0521,-0.0354>
                    ENDMEAS/
                    PNT48 =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                    THEO/<1.0989,4.2011,-0.1411>,<0.9999785,-0.0008633,0.0064954>
                    ACTL/<1.0989,4.2011,-0.1411>,<0.9999785,-0.0008633,0.0064954>
                    TARG/<1.0989,4.2011,-0.1411>,<0.9999785,-0.0008633,0.0064954>
                    SNAP=NO
                    SHOW HITS=YES
                    HIT/BASIC,<1.0989,4.2011,-0.1411>,<0.9999785,-0.0008633,0.0064954>,<1.0989,4.2011,-0.1411>
                    ENDMEAS/
                    SF16 =FEAT/PLANE,CARTESIAN,TRIANGLE,NO
                    THEO/<1.0985,4.0484,-0.1029>,<0.9999785,-0.0008633,0.0064954>
                    ACTL/<1.0985,4.0484,-0.1029>,<0.9999785,-0.0008633,0.0064954>
                    CONSTR/PLANE,BF,PNT46,PNT47,PNT48,,
                    D16 =SURFACE PROFILE : SF16
                    COMPOSITE=NO,
                    CUSTOMIZED DRF=NO
                    NOTE/D16
                    FEATURES/SF16,,
                    PNT49 =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                    THEO/<1.0901,4.3876,-0.0837>,<0.9658879,0.2589039,0.0054117>
                    ACTL/<1.0901,4.3876,-0.0837>,<0.9658879,0.2589039,0.0054117>
                    TARG/<1.0901,4.3876,-0.0837>,<0.9658879,0.2589039,0.0054117>
                    SNAP=NO
                    SHOW HITS=YES
                    HIT/BASIC,<1.0901,4.3876,-0.0837>,<0.9658879,0.2589039,0.0054117>,<1.0901,4. 3876,-0.0837>
                    ENDMEAS/
                    PNT50 =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                    THEO/<1.0243,4.5008,-0.0805>,<0.7027169,0.7114662,0.0022006>
                    ACTL/<1.0243,4.5008,-0.0805>,<0.7027169,0.7114662,0.0022006>
                    TARG/<1.0243,4.5008,-0.0805>,<0.7027169,0.7114662,0.0022006>
                    SNAP=NO
                    SHOW HITS=YES
                    HIT/BASIC,<1.0243,4.5008,-0.0805>,<0.7027169,0.7114662,0.0022006>,<1.0243,4. 5008,-0.0805>
                    ENDMEAS/
                    PNT51 =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                    THEO/<0.9154,4.5638,-0.0871>,<0.2668431,0.9637389,-0.0014668>
                    ACTL/<0.9154,4.5638,-0.0871>,<0.2668431,0.9637389,-0.0014668>
                    TARG/<0.9154,4.5638,-0.0871>,<0.2668431,0.9637389,-0.0014668>
                    SNAP=NO
                    SHOW HITS=YES
                    HIT/BASIC,<0.9154,4.5638,-0.0871>,<0.2668431,0.9637389,-0.0014668>,<0.9154,4.5638,-0.0871>
                    ENDMEAS/
                    SF17 =FEAT/SET,CARTESIAN
                    THEO/<1.0099,4.4841,-0.0838>,<0,0,1>
                    ACTL/<1.0099,4.4841,-0.0838>,<0,0,1>
                    CONSTR/SET,BASIC,PNT49,PNT50,PNT51,,
                    D17 =SURFACE PROFILE : SF17
                    COMPOSITE=NO,
                    CUSTOMIZED DRF=NO
                    NOTE/D17
                    FEATURES/SF17,,
                      report.
                      SAMPLE_003.PDF
                        Why are you using the iterative points to construct planes?
                        sigpicIt's corona time!
                          MOVE/CLEARPLANE
                          SNAP=NO
                          SHOW HITS=YES
                          ENDMEAS/
                          Theo and targets are different
                            For the “A” because I need a datum surface and I have never had good luck using a constructed feature set as a datum. The B & C because they are planes.