  • Probe calibraton - Probe changer

    Hi,

    I'm using hexagon auto probe changer with 4 slot at machine where I using only 4 same programs every day.

    My question is, do I need to calibrate probes for that 4 programs every morning when I'm turn on machine if i have auto probe changer? Probes and sensors are always the same on that machine. Set is SF_1x20, SF_0.5x20, SF_2x20 and LF_0.3x10.

  • #2
    No not really.

    I have 12 probes with many many angles - full calibration takes half a day at least. We run a weekly probe check program (checks a couple of angles from each probe against all the others) and if they all tally we just leave it.

    If any are out then they get re-calibrated.