I may be about to display a total lack of some fundamentals, so please don't be too harsh...
I have the CMM using autocylinder to measure an OD with a callout 0.636-0.637 diameter. The length of the cylinder I can hit with my probe is pretty short, but I put three levels on it, since it is a primary datum. There is no cylindricity callout, but my hits showed lobing, so I started looking at cylindricity, and PCDMIS reports 0.0008.
1) I know the CMM probe can cause lobing in measurements. I have been under the impression that PCDMIS uses calibration data to correct for this lobing when it does all its calculations, before reporting the measurement. Does it?
2) Basic metrology question: my tolerance is basically 0.001 total. With cylindricity 0.0008, that is using up most of the tolerance, and I think my result would have to be within 0.0001 of the mean for the part to still be good. Is this correct?
3) I have the default "least squares" method on. I did try the "minimum circumscribed" on a part to confirm that method gives me a result that is quite a bit bigger. How do y'all decide which to use? (Part of the reason I did this is to show the operators why the CMM is getting different numbers than they are getting with digital mics at the machine. No matter where you put the mic, you're at a high and a low point, so they were ALWAYS getting lower numbers than the least square result and wanted to discount the CMM results. I need to be smart enough to explain what's going on so we can make the best parts we can.)
I keep reading and learning, but often come up against questions that I don't have the knowledge to figure out yet. Thank you in advance for helping out!
JuneScreen.jpg
I may be about to display a total lack of some fundamentals, so please don't be too harsh...
I have the CMM using autocylinder to measure an OD with a callout 0.636-0.637 diameter. The length of the cylinder I can hit with my probe is pretty short, but I put three levels on it, since it is a primary datum. There is no cylindricity callout, but my hits showed lobing, so I started looking at cylindricity, and PCDMIS reports 0.0008.
1) I know the CMM probe can cause lobing in measurements. I have been under the impression that PCDMIS uses calibration data to correct for this lobing when it does all its calculations, before reporting the measurement. Does it?
2) Basic metrology question: my tolerance is basically 0.001 total. With cylindricity 0.0008, that is using up most of the tolerance, and I think my result would have to be within 0.0001 of the mean for the part to still be good. Is this correct?
3) I have the default "least squares" method on. I did try the "minimum circumscribed" on a part to confirm that method gives me a result that is quite a bit bigger. How do y'all decide which to use? (Part of the reason I did this is to show the operators why the CMM is getting different numbers than they are getting with digital mics at the machine. No matter where you put the mic, you're at a high and a low point, so they were ALWAYS getting lower numbers than the least square result and wanted to discount the CMM results. I need to be smart enough to explain what's going on so we can make the best parts we can.)
I keep reading and learning, but often come up against questions that I don't have the knowledge to figure out yet. Thank you in advance for helping out!