  • Probe calibration - Wrong parameter

    Hi all,
    i've got an issue with probe calibration. If i want to start tip cal, pc dmis will write that the parameter is wrong. I tried to reset tips, create new tips, call backup, copy new PROBE.dat. The machine can normally measure, but i cannot start a calibration.

    Thanks for feedbacks.



    Probe_issue.png


    Attached Files

  • #2
    I can barely see it, but you have parameters I've not seen for the Joint angles (the two lines directly below the probe head). Can you double click and change them to zero?