    Just got a new global performance with pc-dmis 2015.1 Setting up a dcc program and when I start the program and it goes to DCC I get an error The error reads:
    position out of the machine stroke limit on Z axis.. What am I doing wrong?
    Alignment? Did you level a plane in "Z" and origin it?
      No, I did not... But I just did... But it stops at the red arrow on the left of the screen... How do I get rid of the red arrow??? Thank You much in advance!!
        Posting the code will help us help you.
        TK
        sigpicHave a homebrew
          Sounds like it's trying to drive to a move point outside the working envelope of the machine. Are you sending it to some large Z+ or Z- location?
          Mike
          You can never tell exactly how deep a mud puddle is without getting your feet wet -- Dennis the Menice
            Mike
            Yep.
            99% of the time, the root cause is an alignment error.
              I'm guessing your new? you need to help the CMM know where the part is on the surface plate of the machine or the next thing it will give you an error in the "X" axis then the "Y". The machine has no idea where the part is on the table.
                Thank You everyone!!! Big help and on track now!!
                  What was your solution?
                  Mike
                    set up a an alignment...
                      Im checking a 123 block with a star probe. Block is on standoffs. When I take points on the bottom of the block go get thickness using tip4 (X-), i get the out of limit error on Z but there is room to move the probe down to. I also use tips 3 and 5 to take hits on the bottom and they work fine, errors out on when asked to move below -1.0 with tip 4 only. If i move the probe down and hit skip the move, it probes the hits, moves over, and goes up as programmed. I re cal'd the tips and also reset the tips and still the same results. Doing 321 alignments for both man and dcc.
                                    perform 3 2 1 alignment
                        PLN1       =FEAT/PLANE,CARTESIAN,TRIANGLE
                                    THEO/<9.5074,12.8389,-13.3207>,<0,0,1>
                                    ACTL/<4.4503,11.9277,-13.2926>,<-0.0069558,-0.0026821,0.9999722>
                                    MEAS/PLANE,3
                                      HIT/BASIC,NORMAL,<8.9189,13.545,-13.3208>,<0,0,1>,<5.0406,12.5239,-13.2869>,USE THEO=YES
                                      HIT/BASIC,NORMAL,<10.0182,13.5495,-13.3217>,<0,0,1>,<4.1544,12.5239,-13.293>,USE THEO=YES
                                      HIT/BASIC,NORMAL,<9.5851,11.4222,-13.3197>,<0,0,1>,<4.1544,10.7348,-13.2978>,USE THEO=YES
                                    ENDMEAS/
                        LIN1       =FEAT/LINE,CARTESIAN,UNBOUNDED
                                    THEO/<9.1398,11.1755,-13.524>,<1,0,0>
                                    ACTL/<4.1543,10.1892,-13.4134>,<0.9999989,0.0015121,0>
                                    MEAS/LINE,2,ZPLUS
                                      HIT/BASIC,NORMAL,<9.1398,11.1755,-13.5245>,<0,-1,0>,<4.1544,10.1892,-13.4134>,USE THEO=YES
                                      HIT/BASIC,NORMAL,<10.1596,11.1755,-13.5236>,<0,-1,0>,<5.0878,10.1907,-13.4134>,USE THEO=YES
                                    ENDMEAS/
                        PNT1       =FEAT/POINT,CARTESIAN
                                    THEO/<8.4864,12.264,-13.6423>,<-1,0,0>
                                    ACTL/<3.5281,11.8135,-13.5677>,<-1,0,0>
                                    MEAS/POINT,1,WORKPLANE
                                      HIT/BASIC,NORMAL,<8.4864,12.264,-13.6423>,<-1,0,0>,<3.5281,11.8135,-13.5677>,USE THEO=YES
                                    ENDMEAS/
                        A1         =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
                                      ALIGNMENT/LEVEL,ZPLUS,PLN1
                                      ALIGNMENT/ROTATE,XPLUS,TO,LIN1,ABOUT,ZPLUS
                                      ALIGNMENT/TRANS,XAXIS,PNT1
                                      ALIGNMENT/TRANS,YAXIS,LIN1
                                      ALIGNMENT/TRANS,ZAXIS,PLN1
                                    ALIGNMENT/END
                                    MODE/DCC
                                    MOVE/POINT,NORMAL,<0,0,1>
                        PLN2       =FEAT/PLANE,CARTESIAN,TRIANGLE
                                    THEO/<0.9395,1.8258,0>,<0,0,1>
                                    ACTL/<0.9394,1.8258,0>,<-0.0000022,0.0000169,1>
                                    MEAS/PLANE,3
                                      HIT/BASIC,NORMAL,<0.4644,2.5215,0>,<0,0,1>,<0.4643,2.5216,0>,USE THEO=YES
                                      HIT/BASIC,NORMAL,<1.4147,2.5213,0>,<0,0,1>,<1.4147,2.5213,0>,USE THEO=YES
                                      HIT/BASIC,NORMAL,<0.9393,0.4344,0>,<0,0,1>,<0.9392,0.4344,0.0001>,USE THEO=YES
                                    ENDMEAS/
                                    MOVE/POINT,NORMAL,<0.6326,-0.447,0.1379>
                        LIN2       =FEAT/LINE,CARTESIAN,UNBOUNDED
                                    THEO/<0.6333,0,-0.3713>,<1,0,0>
                                    ACTL/<0.6333,-0.0001,-0.3713>,<1,-0.0000237,0>
                                    MEAS/LINE,2,ZPLUS
                                      HIT/BASIC,NORMAL,<0.6333,0,-0.3713>,<0,-1,0>,<0.6333,-0.0001,-0.3713>,USE THEO=YES
                                      HIT/BASIC,NORMAL,<1.5014,0,-0.3713>,<0,-1,0>,<1.5014,-0.0001,-0.3713>,USE THEO=YES
                                    ENDMEAS/
                                    MOVE/POINT,NORMAL,<1.5014,-0.2025,1>
                                    MOVE/POINT,NORMAL,<-.50,1.5,1>
                        PNT2       =FEAT/POINT,CARTESIAN
                                    THEO/<0,1.465,-0.1456>,<-1,0,0>
                                    ACTL/<-0.0001,1.465,-0.1456>,<-1,0,0>
                                      HIT/BASIC,NORMAL,<0,1.465,-0.1456>,<-1,0,0>,<-0.0001,1.465,-0.1456>,USE THEO=YES
                                    ENDMEAS/
                        A2         =ALIGNMENT/START,RECALL:A1,LIST=YES
                                      ALIGNMENT/LEVEL,ZPLUS,PLN2
                                      ALIGNMENT/ROTATE,XPLUS,TO,LIN2,ABOUT,ZPLUS
                                      ALIGNMENT/TRANS,XAXIS,PNT2
                                      ALIGNMENT/TRANS,YAXIS,LIN2
                                      ALIGNMENT/TRANS,ZAXIS,PLN2
                                    ALIGNMENT/END
                                    MOVE/POINT,NORMAL,<-0.1951,1.4649,1>
                          top_plane  =feat/plane,cartesian,triangle
                                      theo/<0.9362,1.6056,0>,<0,0,1>
                                      actl/<0.9362,1.6056,0>,<0.0000242,-0.0000214,1>
                                      meas/plane,12
                                        hit/basic,normal,<0.3819,2.609,0.0001>,<0,0,1>,<0.3818,2.609,0>,use theo=yes
                                        hit/basic,normal,<1.0849,2.6093,0>,<0,0,1>,<1.0849,2.6093,0>,use theo=yes
                                        hit/basic,normal,<1.5271,2.6094,0>,<0,0,1>,<1.527,2.6094,-0.0001>,use theo=yes
                                        hit/basic,normal,<1.527,1.8982,0>,<0,0,1>,<1.5269,1.8982,0>,use theo=yes
                                        hit/basic,normal,<0.9054,1.8988,0>,<0,0,1>,<0.9054,1.8988,0>,use theo=yes
                                        hit/basic,normal,<0.2278,1.8989,0>,<0,0,1>,<0.2278,1.8989,0>,use theo=yes
                                        hit/basic,normal,<0.2273,1.3299,0>,<0,0,1>,<0.2273,1.3299,0>,use theo=yes
                                        hit/basic,normal,<0.8599,1.3302,0>,<0,0,1>,<0.8599,1.3302,0>,use theo=yes
                                        hit/basic,normal,<1.5014,1.3302,0>,<0,0,1>,<1.5014,1.3303,0>,use theo=yes
                                        hit/basic,normal,<1.5007,0.5842,0>,<0,0,1>,<1.5007,0.5841,-0.0001>,use theo=yes
                                        hit/basic,normal,<1.0331,0.5843,0>,<0,0,1>,<1.0331,0.5843,-0.0001>,use theo=yes
                                        hit/basic,normal,<0.4586,0.5843,0>,<0,0,1>,<0.4586,0.5843,-0.0001>,use theo=yes
                                      endmeas/
                                      move/point,normal,<0.529,-0.4147,0.199>
                          front_plane=feat/plane,cartesian,triangle
                                      theo/<1.1872,0,-0.3768>,<0,-1,0>
                                      actl/<1.1872,0,-0.3768>,<0.0000202,-1,0.0002009>
                                      meas/plane,12
                                        hit/basic,normal,<0.5294,0.0001,-0.2248>,<0,-1,0>,<0.5295,0,-0.2249>,use theo=yes
                                        hit/basic,normal,<0.8966,0.0001,-0.2248>,<0,-1,0>,<0.8967,0,-0.2248>,use theo=yes
                                        hit/basic,normal,<1.3153,-0.0001,-0.2243>,<0,-1,0>,<1.3153,0.0001,-0.2243>,use theo=yes
                                        hit/basic,normal,<1.8059,0,-0.224>,<0,-1,0>,<1.806,0.0002,-0.224>,use theo=yes
                                        hit/basic,normal,<1.806,-0.0001,-0.3905>,<0,-1,0>,<1.806,-0.0001,-0.3905>,use theo=yes
                                        hit/basic,normal,<1.4511,0,-0.3911>,<0,-1,0>,<1.451,0.0001,-0.3911>,use theo=yes
                                        hit/basic,normal,<1.0317,0,-0.3907>,<0,-1,0>,<1.0317,0,-0.3908>,use theo=yes
                                        hit/basic,normal,<0.6385,-0.0001,-0.391>,<0,-1,0>,<0.6385,0.0001,-0.3911>,use theo=yes
                                        hit/basic,normal,<0.6384,0,-0.5153>,<0,-1,0>,<0.6384,0,-0.5154>,use theo=yes
                                        hit/basic,normal,<1.017,0.0001,-0.5147>,<0,-1,0>,<1.017,0,-0.5147>,use theo=yes
                                        hit/basic,normal,<1.3252,0,-0.5147>,<0,-1,0>,<1.3253,0.0001,-0.5147>,use theo=yes
                                        hit/basic,normal,<1.7913,0.0001,-0.5151>,<0,-1,0>,<1.7913,0,-0.5151>,use theo=yes
                                      endmeas/
                                      move/point,normal,<1.7914,-0.2043,1>
                                      tip/t3a0b0, shankijk=0.0003, -0.9999, -0.0114, angle=0.9464
                                      move/point,normal,<1.5144,-0.7323,1.0444>
                                      move/point,normal,<1.5145,-0.7322,-1.082>
                          pnt3       =feat/point,cartesian
                                      theo/<1.5144,0.0848,-1>,<0,0,-1>
                                      actl/<1.5144,0.0848,-0.9997>,<0,0,-1>
                                      meas/point,1,workplane
                                        hit/basic,normal,<1.5144,0.0848,-1>,<0,0,-1>,<1.5144,0.0848,-0.9997>,use theo=yes
                                      endmeas/
                          pnt4       =feat/point,cartesian
                                      theo/<1.1685,0.0849,-1>,<0,0,-1>
                                      actl/<1.1686,0.0849,-0.9997>,<0,0,-1>
                                        hit/basic,normal,<1.1685,0.0849,-1>,<0,0,-1>,<1.1686,0.0849,-0.9997>,use theo=yes
                                      endmeas/
                          pnt5       =feat/point,cartesian
                                      theo/<0.7079,0.0849,-1>,<0,0,-1>
                                      actl/<0.7079,0.0849,-0.9997>,<0,0,-1>
                                        hit/basic,normal,<0.7079,0.0849,-1>,<0,0,-1>,<0.7079,0.0849,-0.9997>,use theo=yes
                                      endmeas/
                                      move/point,normal,<1.4075,-0.7619,-1.1379>
                                      move/point,normal,<1.4075,-0.7618,1.832>
                                      tip/t1a0b0, shankijk=0.0003, -0.9999, -0.0114, angle=0.9464
                                      move/point,normal,<-.5,0.5,1>
                          leftplane  =feat/plane,cartesian,triangle
                                      theo/<0,1.4658,-0.2923>,<-1,0,0>
                                      actl/<0.0002,1.4658,-0.2923>,<-0.9999999,-0.0000181,-0.000411>
                                      meas/plane,12
                                        hit/basic,normal,<0,0.4995,-0.1477>,<-1,0,0>,<0,0.4995,-0.1477>,use theo=yes
                                        hit/basic,normal,<0,1.0207,-0.1477>,<-1,0,0>,<0.0001,1.0207,-0.1477>,use theo=yes
                                        hit/basic,normal,<0.0001,1.6503,-0.1477>,<-1,0,0>,<0.0002,1.6502,-0.1477>,use theo=yes
                                        hit/basic,normal,<0,2.5053,-0.1477>,<-1,0,0>,<0.0001,2.5052,-0.1477>,use theo=yes
                                        hit/basic,normal,<0,2.5053,-0.2838>,<-1,0,0>,<0.0001,2.5053,-0.2838>,use theo=yes
                                        hit/basic,normal,<-0.0001,1.8919,-0.2839>,<-1,0,0>,<0.0002,1.8919,-0.2839>,use theo=yes
                                        hit/basic,normal,<0,1.0375,-0.2848>,<-1,0,0>,<0.0002,1.0375,-0.2848>,use theo=yes
                                        hit/basic,normal,<0,0.5637,-0.2848>,<-1,0,0>,<0.0002,0.5636,-0.2848>,use theo=yes
                                        hit/basic,normal,<0,0.5637,-0.4449>,<-1,0,0>,<0.0002,0.5636,-0.4449>,use theo=yes
                                        hit/basic,normal,<-0.0001,1.063,-0.445>,<-1,0,0>,<0.0002,1.063,-0.445>,use theo=yes
                                        hit/basic,normal,<0.0002,1.8041,-0.445>,<-1,0,0>,<0.0003,1.804,-0.445>,use theo=yes
                                        hit/basic,normal,<0,2.485,-0.445>,<-1,0,0>,<0.0001,2.485,-0.445>,use theo=yes
                                      endmeas/
                                      move/point,normal,<-0.19,2.4851,1>
                                      move/point,normal,<0.5,3.3,1>
                          back_plane =feat/plane,cartesian,triangle
                                      theo/<1.0039,3,-0.3282>,<0,1,0>
                                      actl/<1.0039,3.0006,-0.3282>,<-0.0000125,0.9999999,-0.0003711>
                                      meas/plane,12
                                        hit/basic,normal,<0.2743,3.0001,-0.2003>,<0,1,0>,<0.2743,3.0007,-0.2002>,use theo=yes
                                        hit/basic,normal,<0.7734,3,-0.2003>,<0,1,0>,<0.7734,3.0007,-0.2002>,use theo=yes
                                        hit/basic,normal,<1.2929,2.9999,-0.2009>,<0,1,0>,<1.2929,3.0006,-0.2008>,use theo=yes
                                        hit/basic,normal,<1.7009,3,-0.2009>,<0,1,0>,<1.7009,3.0007,-0.2009>,use theo=yes
                                        hit/basic,normal,<1.7004,3,-0.331>,<0,1,0>,<1.7004,3.0006,-0.331>,use theo=yes
                                        hit/basic,normal,<1.2599,2.9999,-0.3304>,<0,1,0>,<1.2599,3.0006,-0.3304>,use theo=yes
                                        hit/basic,normal,<0.7471,3,-0.3293>,<0,1,0>,<0.7471,3.0006,-0.3293>,use theo=yes
                                        hit/basic,normal,<0.3162,3,-0.3293>,<0,1,0>,<0.3161,3.0005,-0.3293>,use theo=yes
                                        hit/basic,normal,<0.3161,2.9999,-0.4538>,<0,1,0>,<0.3161,3.0006,-0.4538>,use theo=yes
                                        hit/basic,normal,<0.8032,3.0001,-0.454>,<0,1,0>,<0.8032,3.0005,-0.4539>,use theo=yes
                                        hit/basic,normal,<1.2396,3.0001,-0.4539>,<0,1,0>,<1.2397,3.0006,-0.4539>,use theo=yes
                                        hit/basic,normal,<1.6229,3,-0.4546>,<0,1,0>,<1.6229,3.0006,-0.4546>,use theo=yes
                                      endmeas/
                            move/point,normal,<1.5,5,0.25>
                                        tip/t5a0b0, shankijk=-0.0003, 0.9999, 0.0114, angle=-0.9464
                                        move/point,normal,<1.5248,4.0748,-1.3566>
                            pnt6       =feat/point,cartesian
                                        theo/<1.6982,2.912,-1>,<0,0,-1>
                                        actl/<1.6983,2.912,-1.0001>,<0,0,-1>
                                          hit/basic,normal,<1.6982,2.912,-1>,<0,0,-1>,<1.6983,2.912,-1.0001>,use theo=yes
                                        endmeas/
                            pnt7       =feat/point,cartesian
                                        theo/<1.2262,2.912,-1>,<0,0,-1>
                                        actl/<1.2262,2.912,-1.0002>,<0,0,-1>
                                          hit/basic,normal,<1.2262,2.912,-1>,<0,0,-1>,<1.2262,2.912,-1.0002>,use theo=yes
                                        endmeas/
                            pnt8       =feat/point,cartesian
                                        theo/<0.7242,2.912,-1>,<0,0,-1>
                                        actl/<0.7243,2.912,-1.0002>,<0,0,-1>
                                          hit/basic,normal,<0.7242,2.912,-1>,<0,0,-1>,<0.7243,2.912,-1.0002>,use theo=yes
                                        endmeas/
                                        move/point,normal,<0.7242,3.5,-1.0787>
                                        move/point,normal,<0.7242,3.5,1.5>
                                        move/point,normal,<2.5608,2.3649,0.465>
                            right_plane=feat/plane,cartesian,triangle
                                        theo/<2,1.296,-0.3831>,<1,0,0>
                                        actl/<2.0004,1.296,-0.3832>,<1,0.0000036,0.0002283>
                                        meas/plane,12
                                          hit/basic,normal,<1.9999,2.3051,-0.1285>,<1,0,0>,<2.0003,2.3051,-0.1286>,use theo=yes
                                          hit/basic,normal,<2,1.6063,-0.1285>,<1,0,0>,<2.0003,1.6063,-0.1286>,use theo=yes
                                          hit/basic,normal,<2,1.0221,-0.1285>,<1,0,0>,<2.0003,1.0221,-0.1285>,use theo=yes
                                          hit/basic,normal,<2,0.286,-0.1285>,<1,0,0>,<2.0004,0.286,-0.1286>,use theo=yes
                                          hit/basic,normal,<2.0001,0.2679,-0.4072>,<1,0,0>,<2.0005,0.2679,-0.4072>,use theo=yes
                                          hit/basic,normal,<2,0.879,-0.4074>,<1,0,0>,<2.0004,0.879,-0.4074>,use theo=yes
                                          hit/basic,normal,<1.9999,1.4151,-0.4073>,<1,0,0>,<2.0003,1.4152,-0.4073>,use theo=yes
                                          hit/basic,normal,<2,2.3339,-0.4069>,<1,0,0>,<2.0004,2.3339,-0.4069>,use theo=yes
                                          hit/basic,normal,<2.0001,2.3337,-0.6137>,<1,0,0>,<2.0004,2.3338,-0.6137>,use theo=yes
                                          hit/basic,normal,<2,1.7747,-0.6138>,<1,0,0>,<2.0004,1.7748,-0.6138>,use theo=yes
                                          hit/basic,normal,<1.9999,1.0314,-0.6137>,<1,0,0>,<2.0004,1.0315,-0.6137>,use theo=yes
                                          hit/basic,normal,<2,0.2964,-0.6137>,<1,0,0>,<2.0003,0.2964,-0.6138>,use theo=yes
                                        endmeas/
                                        move/point,normal,<3.1401,0.2963,-0.6137>
                                        tip/t4a0b0, shankijk=0.9999, 0.0001, 0.0165, angle=0.6549
                                        comment/rept,
                                        this is where i get the error
                                        if i manually move down in z and skip the move
                                        it proceeds to probe the 3 points
                                        the the program continues
                                        move/point,normal,<2.2195,0.2855,-1.3822>
                            pnt9       =feat/point,cartesian
                                        theo/<1.908,0.7288,-1>,<0,0,-1>
                                        actl/<1.9081,0.729,-1.0002>,<0.0000129,0.0000029,-1>
                                          hit/basic,normal,<1.9081,0.7288,-1>,<0,0,-1>,<1.9081,0.729,-1.0002>,use theo=yes
                                        endmeas/
                            pnt10      =feat/point,cartesian
                                        theo/<1.908,1.3186,-1>,<0,0,-1>
                                        actl/<1.908,1.3189,-1.0003>,<0.0000129,0.0000029,-1>
                                          hit/basic,normal,<1.908,1.3187,-1>,<0,0,-1>,<1.908,1.3189,-1.0002>,use theo=yes
                                        endmeas/
                            pnt11      =feat/point,cartesian
                                        theo/<1.9086,2.1807,-1>,<0,0,-1>
                                        actl/<1.9086,2.1809,-1.0002>,<0.0000193,-0.000003,-1>
                                          hit/basic,normal,<1.9086,2.1807,-1>,<0,0,-1>,<1.9086,2.1809,-1.0002>,use theo=yes
                                        endmeas/
                                        move/point,normal,<2.5601,2.1808,-1.2293>
                                        move/point,normal,<2.5601,0.7074,2.7797>
                                        move/point,normal,<1.63,5.5211,5>
                                        movespeed/ 80
                            bottom_plane=feat/plane,cartesian,triangle,yes
                                        theo/<1.4182,1.4688,-1>,<0,0,-1>
                                        actl/<1.4183,1.4689,-1>,<0,0.0000124,-1>
                                        constr/plane,bfre,pnt3,pnt4,pnt5,pnt6,pnt7,pnt8,pnt9,pnt10,pnt11,,
                                        outlier_removal/off,3
                                        filter/off,wavelength=0
                                        workplane/xplus
                              Change Z (before and after) to -1.08 and try. I suspect the move is low close to the granite?