I'm having problems calibrating the subject stylus using PC-DMIS 4.1 with an SP600. I keep getting an error stating that the defection matrix couldn't be calculated. I have used this setup in earlier version of PC-DMIS. Has something changed in the software?
Carbon Fiber 8mm by 300mm
Collapse
X
-
Brice are you trying to calibrate your lower matrix with 300mm of length of extension between sp600 and your head, or do you have 300mm on the end of you sp600. Either way that is to long. Go with the 8mmx50 that should have come with the sp600 to do the lower matrix. Also I got the same error message when I did my lower matrix, and I was told to ignore it. (see your other post)Time for the Trolls to leave.
-
I think it has something to do with the Opt.Probe parameters, edit>preferences>parameters. I found some documentation from B&S about the Lower Matrix, in the document it had different parameter. I changed to the ones in the document and it made things worse. It seems like B&S should have some kind of document explaining the influence of the parameters on the software. The SP600 works well with styli that are 100mm or less. I thought I saw on the renishaw website that the SP600 could carry up to 300mm with the extended module. I now see that they are saying 280mm/20grams. Maybe I'm outside the envelope, but I know I used this stylus before with good results.
Any thoughts on the parameters?
Comment
Related Topics
Collapse
-
by brice1Is there documentation on the SP600 lower matrix, the only information I found was on the wilcox website.
If one had a perfect SP600 what...-
Channel: PC-DMIS for CMMs
12-15-2006, 09:55 AM -
-
by Ironhoethrow a ceramic 8x50 on the thing, clean everything spotless and let 'er buck... Im guessing you have some dirt, deflection, loose connection, something...
-
Channel: PC-DMIS for CMMs
03-18-2010, 11:25 AM -
-
Has anyone encountered an issue where the probe changer need to be re calibrated after performing a lower matrix calibration? Also, I am using my S_5x50...
-
Channel: PC-DMIS for CMMs
05-06-2019, 10:30 AM -
-
by giantslayerCan anyone assist me with directions on how to perform a lower matrix? Hexagon did one when we replaced our SP600 and our PH10M but I have been un successful...
-
Channel: PC-DMIS for CMMs
02-24-2015, 01:35 PM -
-
by Chally72Anyone know how to display the lower-level matrix from the controller in hyperterminal? Thanks!
-
Channel: PC-DMIS for CMMs
11-24-2010, 05:56 AM -
Comment