Tried to search, no luck. I've only ever calibrated TTP star probes on straight up artifacts, however with the SP25 its not allowing me to set a start and end angle on my calibration. Is the only way to calibrate this thing to use parameter sets and multiple angles on the calibration sphere?
SP25+Star Probe Calibration?
Collapse
X
-
Unfortunately yes, for the full calibration that has to be done first. The calibration routine for full calibration is locked. Something else that can help is to select DCC+DCC and use a shorter prehit and retraction distance. I've found this can help with close calls during calibration.
That being said, once it's been done you can do a partial calibration which gives you the standard definitions of number of hits and start & end angles. The partial will only work once the full scanning calibration has been done and the probe build hasn't been broken down. Even a crash can cause the partial calibration to give less than desirable results.PC-DMIS 2016.0 SP8
Jeff
-
Originally posted by Josh Carpenter View PostI'm guessing that your qual sphere only has a 45 degree base, right?
Try and craft some sort of adapter (I've used sine plates in the past) to get the qual sphere pointing straight up, and make a new tool definition with IJK 0,0,1.
Then your star tips should run.
Comment
-
Originally posted by fineout12 View PostI have one of the renishaw spheres that can rotate to 45 degrees but is usually at 001, but when it goes to measure the sides of the star at 001 it is hitting the shank of the sphere.
Comment
-
Originally posted by Josh Carpenter View PostPlease describe the wrist angle and which way the actual tips are pointing when this happens.Last edited by fineout12; 12-08-2015, 04:09 PM.
Comment
-
Originally posted by fineout12 View PostT1A0B0 goes fine T2A0B0 crashes. Pc-Dmis 4.3 MR1 CAD++ 5 way star with equal length tips are fairly square to the machine coordinates.
Make sure qual sphere tool definition has IJK 0,0,1.
Fairly square is not square enough. You need to run tips back and forth near a block, eyeballing until they are VERY square to machine travel.
Just checking: you are using a master probe (preferably a short stiff large dia ball) in A0 B0 to locate qual sphere with manual hit, not the star probe, right?
If it still does not work, time to go to Plan B:
1) 4 New qual tool definitions, each mounted parallel to granite with IJKs to match. Each one is located by master probe A0 B0, then calibrate the star tip that points at that tool orientation.
2) Go back to sphere IJK 0,0,1 and do the quick cal check.
Comment
-
Originally posted by Josh Carpenter View PostCheck that the virtual build matches real-life build EXACTLY.
Make sure qual sphere tool definition has IJK 0,0,1.
Fairly square is not square enough. You need to run tips back and forth near a block, eyeballing until they are VERY square to machine travel.
Just checking: you are using a master probe (preferably a short stiff large dia ball) in A0 B0 to locate qual sphere with manual hit, not the star probe, right?
If it still does not work, time to go to Plan B:
1) 4 New qual tool definitions, each mounted parallel to granite with IJKs to match. Each one is located by master probe A0 B0, then calibrate the star tip that points at that tool orientation.
2) Go back to sphere IJK 0,0,1 and do the quick cal check.
Comment
-
I've attached some information from Renishaw on star probe configurations, I don't know if it will helpAttached Files
Comment
Related Topics
Collapse
-
by FUKNWATAnyone know why anyone would need an angled calibration sphere? My company has one and I want to replace it with a regular vertical calibration spher...
-
Channel: PC-DMIS for CMMs
03-25-2016, 01:47 PM -
-
by Ron OLoane
I am running v3.7 mr3
Ocasionally when doing a probe calibration on a program that has been executed 100's of times, the calibration...-
Channel: PC-DMIS for CMMs
05-10-2006, 01:13 PM -
-
by pab39Hi all,
We have an Optiv 321 GL and decided to send the "sphere+ring" calibration tool for calibration every 5 years, and here...-
Channel: PC-DMIS for CMMs
05-14-2019, 03:17 PM -
-
by hammerI am having trouble with star probe calibration I am only using 2 tips locations 1 and 5 and using all angles b-90 and b90 so I have to use 2 calibration...
-
Channel: PC-DMIS for CMMs
10-25-2007, 04:27 PM -
-
by BorisI have a question about probe star calibration
When I made calibration of star probe everithing going fine; but when i going to check...-
Channel: PC-DMIS for CMMs
05-12-2006, 04:29 PM -
Comment