We have been using a TP-200 with out RS-670 running 3.7 MR3. The probe has pretty much reached the end of its lifespan and we attempted to replace it with a more appropriate 200b due to the fact that we have the cmm on an open floor with other machines causing some vibration. The issue is that I am unable to calibrate using the 200b. It reads a stylus diameter of .0778 while I run a 2bx30mm (.08) probe. Right now the theoretical diameter is .0787 and I am unsure how to change that globally. When measuring my calibration sphere with several probe angles my deviation from theoretical is identical so it is still very consistent.
RS-670 TP200b problems
Collapse
X
-
Is it possible that what you are seeing is the "effective" diameter of the calibrated tip. It does not seem to be far off from the actual diameter.
The only other thing is to double check that your master ball diameter is set correctly.
-
I double checked the calibration sphere and its set at the correct size. Also the TP-200 using the same stylus and module pass qualification with no issues on the same ball. Repeatability is great with the 200b but it will not save probe calibration data unless the diameter is closer to theoretical. It seems as if either pcdmis or the probe interface is unable to compensate for the new deflection rate of the less sensitive probe.
Comment
-
I havent had much time to play with the 200B due to mu current workload but I am no closer to a solution to my issue. During calibration i find that every styli i try reads .001 smaller than actual and will fail calibration. Might this be a 3.7 MR3 issue? Anyone else around running a 200b by any chance?
Comment
-
I use a tp200. I considered a tp200b but my sales rep told me every one of those he has sold he ended up exchanging for tp200! FYI
I would not be too concerned about the that the size reads .001" small if it is consistant, and you can run the calibration routine as "qualification check" and it reports your sphere diameter correctly. If it is repeating and giving accurate readings, then like Craig suggested, up the tol in the calibration routine to accept what you are consistantly getting. HTH
sigpic"Hated by Many, Loved by Few" _ A.B. - Stone brewery
Comment
-
Originally posted by bob mappesjust for kicks....try putting a mic on the stylus ball....maybe it IS .0778 in dia.
bob
Comment
Related Topics
Collapse
-
by kuulpinoyHow do I fix measured probe diameter error exceeds the limit?
-
Channel: PC-DMIS for CMMs
07-15-2014, 07:42 PM -
-
by BG619I have a 2mm tip that is measuring as 1.789mm, I tried to correct the probe diameter in the edit window ( within the calibration window) and each time...
-
Channel: PC-DMIS for CMMs
10-25-2021, 05:26 PM -
-
by AW521We have the machine set up and calibrated by a representative from Hexagon. After he left, the machine has been consistently off by 0.0007" on the...
-
Channel: PC-DMIS for CMMs
08-11-2016, 03:25 PM -
-
by MrCommentHey Guys,
Still working out the bugs on my Probe Qualification blues.
We got the StDev better because there was something...-
Channel: PC-DMIS for CMMs
08-01-2007, 12:40 PM -
-
When I qualify the probe tip it measures .0015 smaller. It doesn't matter what the probe diameter is 1,2 or 3 mm it will say probe exceeds deviation....
-
Channel: PC-DMIS for CMMs
07-26-2011, 12:05 PM -
Comment