After reading the responses to the thread for "Checking Clearance of Diameter" and seeing the consistency in those response, I need to know why? If the machine is supposed to be accurate within 3.1 microns, why can't it check .0002"?
I have a Global Image 555 calibrated to ISO-10360-2. The calibration certificate shows:
Maximum Permissible Maximum Observed Error
Error Specification (um) (Actual Recordings at Customer)
Length Meas 1.7 + L/K 3.1 at 460 3.1 @ 460
(E A + L/K)
Probing Test 1.9 1.8
(R rmax - rmin)
Some of the tolerances on the dimensions being measured are +/- .0002" (mostly true position). Is this machine NOT the right one for this application?
Thanks,
Keith