I have fought it tooth and nail but the Quality Manager tells me I have to do Software/Process Validation for PCDMIS. I was under the impression that I could take PCDMIS out of the box and start using it. Also, the annual Calibrations from B&S should tell us that the CMM is doing what its suppose to. Anyway, he not only wants PCDMIS validated but he also wants all of my individual CMM programs validated......kinda like I'm writing a software program instead of a CMM program......bottom line, its just another knee jerk reaction to getting certified to ISO:13485 for medical devices
In my mind, I just want to know that PCDMIS is giving me the true results from my measurements........Plus, I want to keep it Simple (K.I.S.S)
So, here is what I want to do.......Take several known standards (Class 1 GageBlocks) and measure them. If the Software tells me that a 1" gageblock checks 1.000005", then I know its giving me correct results.......Software VALIDATED!!!!!!...........
Do you think this would be acceptable to an auditor?
Thanks,
ZeissUser
In my mind, I just want to know that PCDMIS is giving me the true results from my measurements........Plus, I want to keep it Simple (K.I.S.S)
So, here is what I want to do.......Take several known standards (Class 1 GageBlocks) and measure them. If the Software tells me that a 1" gageblock checks 1.000005", then I know its giving me correct results.......Software VALIDATED!!!!!!...........
Do you think this would be acceptable to an auditor?
Thanks,
ZeissUser