Automated tips program?
  • Automated tips program?
    Hi everyone! I need your help if possible.
    My shop used MM4 before I worked here and there is a program for tip calibration. My boss is looking for something like this for PC-Dmis. I do not think pcd has such a thing, and I didn't find much in favor when I searched the forum.
    Does anyone know of a way to calibrate tips via a program? We are really looking for less operator interaction. Any ideas?
    TIA,
    Becky
    BeckyK
    Sheffield, B&S
    (1)1808-M, (4)D-28, (1)RS-70, (3)RS-150, (1)Mxcel PFX
    PC-Dmis 2009 MR1, 2013 MR1
    TIA,
    Becky
    LOADPROBE/???????
    MOVESPEED/ ???
    TOUCHSPEED/ 1
    COMMENT/DOC,NO,calibrate
    C1 =COMMENT/YESNO,YES,DOYOU NEED TO MEASURE THE BALL POSITION?
    IF_GOTO/C1.INPUT=="NO",GOTO = DCC MEAS
    MODE/DCC
    MOVE/POINT,NORMAL,5,390,-175
    AUTOCALIBRATE/PROBE, PARAMETER_SET=ALL-TIPS-WITH-DEFAULTS, QUALTOOL_MOVED=YES,
    SHOW_SUMMARY=NO, OVERWRITE_RESULTSFILE=YES
    MOVE/POINT,NORMAL,160,-20,-225
    DCC MEAS =LABEL/
    MODE/DCC
    AUTOCALIBRATE/PROBE, PARAMETER_SET=ALL-TIPS-WITH-DEFAULTS, QUALTOOL_MOVED=NO,
    This will work.
    Edity the move points to suit your m/c.
    Roses are red
    Violets are glorious
    Don't try to sneak up
    On Oscar Pistorius.
    MIL-TFP-41C
    3.7mr3, 2012mr1, 2013mr1
      Sure -
      You write a program to do that.
      Don't remember the exact commands (not at the CMM at the moment) -
      After the LOAD PROBE command, insert AUTOCALIBRATE. Do this for every
      probe file. Depending, you may need MOVE POINTS in between, and you may
      need to have the probes to calibrate in a particular order. Set up PARAMETER SETS in the probe dialougue box in the MEASURE window.
      Have a meeting to attend - gotta go!
        If you have a rack with multiple probes make sure you cal the master tip first with the "YES tool moved" setting.
          PART NAME  : Calibration Program
          REV NUMBER : Spectrum Housing
          SER NUMBER : All Operations
          STATS COUNT : 1
          STARTUP    =ALIGNMENT/START,RECALL:, LIST= YES
                      ALIGNMENT/END
                      MODE/MANUAL
                      ACCELERATION/MAXACCELX=2125, MAXACCELY=2125, MAXACCELZ=2125
                      MOVESPEED/ 50
                      TOUCHSPEED/ 1
                      FORMAT/TEXT,OPTIONS,ID,HEADINGS,SYMBOLS, ;NOM,TOL,MEAS,MAXMIN,DEV,DEVANG,OUTTOL
                      COMMENT/OPER,NO,Check probe configuration
                                     ,Slot 1 = 4X50 Probe (Master Probe)
                                     ,Slot 2 = 2X20 W/10EXT
                                     ,Slot 3 = 4X70  to start
                                     ,& will swap the
                                     ,               1X90 90 deg 1 mm ball
                                     ,               t5 - 1X10
                                     ,               t1 thru t4 are empty
                                     ,for it mid program
                                     ,Slot 4 = Star3 t1 - 2X20
                                     ,               t2 - dummy
                                     ,               t3 - dummy
                                     ,               t4 - 4X20
                                     ,               t5 - 2X20
                                     ,Slot 5 - 3 X 50
                                     ,& will swap the
                                     ,               Disk
                                     ,for it mid program & then the 1 X 40
                                     ,Slot 6 - 1 X 20
                                     ,Select "OK" when ready.
                      COMMENT/OPER,NO,Remove part fixture.
                                     ,Install Calibration Tool in hole 12,L,
                                     ,pointing 90 degrees, up, in the "Z" direction
                      COMMENT/DOC,NO,----------------------------------------------------------------------
                                    ,Find Calibration Sphere with Master Probe (4X50)
                      MODE/DCC
                      MOVESPEED/ 50
                      TOUCHSPEED/ 1
                      COMMENT/OPER,NO,CMM is about to load probe from station 1!!!!!
                      LOADPROBE/4X50
                      FORMAT/TEXT,OPTIONS, ,HEADINGS,SYMBOLS, ;MEAS,NOM,DEV,TOL,OUTTOL, ,
                      AUTOCALIBRATE/PROBE, PARAMETER_SET=Master Z Plus Tool Probe, QUALTOOL_MOVED=YES,
                                    SHOW_SUMMARY=YES, OVERWRITE_RESULTSFILE=YES
          BEGIN AUTOCALIBRATE RESULTS FOR PROBE 4X50 USING SET Master Z Plus Tool Probe
          Probe file=4X50    Date=9/4/2008    Time=9:38:22 PM
          tool1          CENT X   3.4928 Y   7.4534 Z -11.0225 D 0.9840
          T1A0B0         THEO X   0.0000 Y   0.0000 Z   9.5098 D 0.1575
          T1A0B0         MEAS X   0.0187 Y  -0.0286 Z   9.5836 D 0.1572 StdDev 0.0000
          END AUTOCALIBRATE RESULTS FOR PROBE 4X50 USING SET Master Z Plus Tool Probe
                                    ,3 X 50 Probe  Calibration
                      LOADPROBE/3 X 50
                      AUTOCALIBRATE/PROBE, PARAMETER_SET=op 20_30, QUALTOOL_MOVED=NO,
          BEGIN AUTOCALIBRATE RESULTS FOR PROBE 3 X 50 USING SET op 20_30
          Probe file=3 X 50    Date=9/4/2008    Time=9:39:08 PM
          T1A0B0         THEO X   0.0000 Y   0.0000 Z   9.5098 D 0.1181
          T1A0B0         MEAS X   0.0365 Y  -0.0409 Z   9.5630 D 0.1177 StdDev 0.0000
          END AUTOCALIBRATE RESULTS FOR PROBE 3 X 50 USING SET op 20_30
                                    ,4 X 50 W/20EXT Probe Calibration
                      LOADPROBE/4X70
          BEGIN AUTOCALIBRATE RESULTS FOR PROBE 4X70 USING SET op 20_30
          Probe file=4X70    Date=9/4/2008    Time=9:41:09 PM
          T1A0B0         THEO X   0.0000 Y   0.0000 Z  10.2972 D 0.1575
          T1A0B0         MEAS X   0.0399 Y  -0.0422 Z  10.3703 D 0.1571 StdDev 0.0000
          T1A90B-90      THEO X   7.2047 Y   0.0000 Z   3.0925 D 0.1575
          T1A90B-90      MEAS X   7.1671 Y  -0.0422 Z   3.2038 D 0.1572 StdDev 0.0000
          T1A45B-90      THEO X   5.0945 Y   0.0000 Z   8.1870 D 0.1575
          T1A45B-90      MEAS X   5.0751 Y  -0.0502 Z   8.2674 D 0.1572 StdDev 0.0000
          END AUTOCALIBRATE RESULTS FOR PROBE 4X70 USING SET op 20_30
                                    ,4X50 Probe Calibration
                      LOADPROBE/4X50
          BEGIN AUTOCALIBRATE RESULTS FOR PROBE 4X50 USING SET op 20_30
          Probe file=4X50    Date=9/4/2008    Time=9:57:12 PM
          T1A7.5B-90     THEO X   0.8376 Y   0.0000 Z   9.4549 D 0.1575
          T1A7.5B-90     MEAS X   0.8571 Y  -0.0334 Z   9.5265 D 0.1572 StdDev 0.0000
          T1A0B0         MEAS X   0.0186 Y  -0.0285 Z   9.5836 D 0.1572 StdDev 0.0000
          T1A7.5B90      THEO X  -0.8376 Y   0.0000 Z   9.4549 D 0.1575
          T1A7.5B90      MEAS X  -0.8181 Y  -0.0347 Z   9.5302 D 0.1572 StdDev 0.0000
          T1A7.5B180     THEO X  -0.0000 Y  -0.8376 Z   9.4549 D 0.1575
          T1A7.5B180     MEAS X   0.0190 Y  -0.8712 Z   9.5283 D 0.1572 StdDev 0.0000
          T1A15B-90      THEO X   1.6609 Y   0.0000 Z   9.2912 D 0.1575
          T1A15B-90      MEAS X   1.6745 Y  -0.0332 Z   9.3614 D 0.1572 StdDev 0.0000
          T1A15B0        THEO X   0.0000 Y   1.6609 Z   9.2912 D 0.1575
          T1A15B0        MEAS X   0.0173 Y   1.6219 Z   9.3649 D 0.1572 StdDev 0.0000
          T1A15B90       THEO X  -1.6609 Y   0.0000 Z   9.2912 D 0.1575
          T1A15B90       MEAS X  -1.6361 Y  -0.0348 Z   9.3688 D 0.1572 StdDev 0.0000
          T1A15B180      THEO X  -0.0000 Y  -1.6609 Z   9.2912 D 0.1575
          T1A15B180      MEAS X   0.0187 Y  -1.6889 Z   9.3650 D 0.1572 StdDev 0.0000
          T1A22.5B0      THEO X   0.0000 Y   2.4558 Z   9.0214 D 0.1575
          T1A22.5B0      MEAS X   0.0148 Y   2.4111 Z   9.0963 D 0.1572 StdDev 0.0000
          T1A30B-180     THEO X   0.0000 Y  -3.2087 Z   8.6501 D 0.1575
          T1A30B-180     MEAS X   0.0191 Y  -3.2258 Z   8.7268 D 0.1572 StdDev 0.0000
          T1A37.5B-15    THEO X   1.0111 Y   3.7735 Z   8.1837 D 0.1575
          T1A37.5B-15    MEAS X   1.0197 Y   3.7193 Z   8.2616 D 0.1572 StdDev 0.0000
          T1A37.5B0      THEO X   0.0000 Y   3.9066 Z   8.1837 D 0.1575
          T1A37.5B0      MEAS X   0.0136 Y   3.8516 Z   8.2633 D 0.1572 StdDev 0.0000
          T1A37.5B15     THEO X  -1.0111 Y   3.7735 Z   8.1837 D 0.1575
          T1A37.5B15     MEAS X  -0.9905 Y   3.7184 Z   8.2657 D 0.1572 StdDev 0.0000
          T1A45B-90      THEO X   4.5377 Y   0.0000 Z   7.6303 D 0.1575
          T1A45B-90      MEAS X   4.5272 Y  -0.0311 Z   7.7028 D 0.1572 StdDev 0.0000
          T1A45B0        THEO X   0.0000 Y   4.5377 Z   7.6303 D 0.1575
          T1A45B0        MEAS X   0.0116 Y   4.4782 Z   7.7128 D 0.1572 StdDev 0.0000
          T1A45B150      THEO X  -2.2689 Y  -3.9298 Z   7.6303 D 0.1575
          T1A45B150      MEAS X  -2.2377 Y  -3.9422 Z   7.7178 D 0.1572 StdDev 0.0000
          T1A52.5B-180   THEO X   0.0000 Y  -5.0912 Z   6.9991 D 0.1575
          T1A52.5B-180   MEAS X   0.0186 Y  -5.0944 Z   7.0852 D 0.1572 StdDev 0.0000
          T1A67.5B-180   THEO X   0.0000 Y  -5.9288 Z   5.5483 D 0.1575
          T1A67.5B-180   MEAS X   0.0170 Y  -5.9252 Z   5.6437 D 0.1572 StdDev 0.0000
          T1A82.5B-90    THEO X   6.3624 Y   0.0000 Z   3.9301 D 0.1575
          T1A82.5B-90    MEAS X   6.3283 Y  -0.0246 Z   4.0228 D 0.1572 StdDev 0.0000
          T1A82.5B-82.5  THEO X   6.3080 Y   0.8305 Z   3.9301 D 0.1575
          T1A82.5B-82.5  MEAS X   6.2732 Y   0.8004 Z   4.0225 D 0.1572 StdDev 0.0000
          T1A82.5B90     THEO X  -6.3624 Y   0.0000 Z   3.9301 D 0.1575
          T1A82.5B90     MEAS X  -6.3135 Y  -0.0427 Z   4.0509 D 0.1572 StdDev 0.0000
          T1A90B-97.5    THEO X   6.3624 Y  -0.8376 Z   3.0925 D 0.1575
          T1A90B-97.5    MEAS X   6.3270 Y  -0.8564 Z   3.1925 D 0.1572 StdDev 0.0000
          T1A90B-180     THEO X   0.0000 Y  -6.4173 Z   3.0925 D 0.1575
          T1A90B-180     MEAS X   0.0153 Y  -6.4085 Z   3.2056 D 0.1572 StdDev 0.0000
          END AUTOCALIBRATE RESULTS FOR PROBE 4X50 USING SET op 20_30
          this is just a part of our cal program
          B&S Global 544
          Using 3.7mr3
          Peace
          Greg
            OK, Stupid question.
            Which one is the "master tip"?
              The probe in slot #1 is the "master" tip.
                and i believe it should be something like a 2x20 probe and not some elaborate build.....
                  TIA,
                  Becky
                  Here is the program I use to calibrate our CMM's:
                  PART NAME  : CALIBRATION
                  REV NUMBER : 1
                  SER NUMBER :
                  STATS COUNT : 1
                  STARTUP    =ALIGNMENT/START,RECALL:,LIST=YES
                              ALIGNMENT/END
                              MODE/DCC
                              COMMENT/OPER,NO,Full Screen=NO,PUT CALIBRATION BALL IN HOLE EA29
                              MOVESPEED/ 50
                              OPTIONPROBE/MAXFORCE=0.15,LOWFORCE=0.025,UPPERFORCE=0.075,TRIGGERFORCE=0.05,
                                PROBEACCURACY=0.1,POSACCURACY=0.1,#RETURNDATA=7,
                                RETURNSPEED=1,SCANPNTDENSITY=4,
                                SCANACCELERATION=10,SCANOFFSETFORCE=0.05,
                                PROBINGMODE=DFL,MANFINEPROBING=NO
                              LOADPROBE/2X20S
                              RECALL/ALIGNMENT,EXTERNAL,a2,PLATE:a2
                              MOVE/POINT,NORMAL,<13.9772,13.9805,7.7952>
                              COMMENT/OPER,NO,Full Screen=NO,TAKE POINT ON TOP OF BALL
                                            SHOW_SUMMARY=NO, OVERWRITE_RESULTSFILE=NO
                              LOADPROBE/1_5X30S
                              TIP/T1A0B0, SHANKIJK=-0.0001, 0.0002, 1, ANGLE=-0.0001
                              LOADPROBE/2X40S
                              LOADPROBE/3X40S
                              LOADPROBE/6X50S
                              LOADPROBE/STAR
                              TIP/T1A0B0, SHANKIJK=-0.0001, 0.0002, 1, ANGLE=0.0589
                              LOADPROBE/5X50S
                              LOADPROBE/6X75S
                              LOADPROBE/18X12S
                              MOVESPEED/ 35
                              LOADPROBE/35X5S
                              LOADPROBE/1X27_5
                              MOVESPEED/ 50
                              LOADPROBE/2X20S
                  DeWain Hodge
                    Yep
                    You are right. A 2x20 is the tip recommended to certify the probe rack. And I keep it in slot one on all CMM's with a rack.
                    Thanks everyone!
                    BeckyK
                    Sheffield, B&S
                      Becky,
                      MM4 had a seperate program because it required you to calibrate the probe angles prior to calling that probefile into the program. If you were programming, and you needed another angle, you had to exit the part program, add the angle, calibrate and go back in. PC-DMIS has this built in. You can calibrate the probe at anytime once a program is open. If you are using the probefile in multiple programs, Ctrl-ALT-P will bring up the probe utility window, and you click "MARK USED". It will only highlight the probe angles your program uses. This eliminates the need to qualify all the tips in the file. However, if your not using A0B0 in the program, I would do that one first, then do the others. The other examples (AUTO/CALIBRATE) given in this thread will work for those people who don't know how to do this, and makes it idiot proof for those non cmm operators.
                        Originally posted by Kenny G View Post
                        Becky,
                        Thanks Kenny, but the issue is that they want a program for the 3rd shift techs to run once a week. Something we can put up on an STI screen. (I guess I should mention that our operators do not actually see PC-Dmis, and they can't have access to it.) I do as you suggest, but what do I know?
                        BeckyK
                        Sheffield, B&S
                          As former MM4 (and AVAIL) programmer, I remember the nightmare of having to calibrate all of the different probe files. The AutoQual routine in pc-dmis allows you to qual as many parameter sets of probes as you want. Each parameter set can have the order user-defined, so there should be no issues with probes smacking anything.
                          And, yes, I know how to "Mark Used" to qual tip angles in a program. (This is a handy tool for those who don't know how to use AutoQual.) That only gets you the tips in that prog.
                          v2013 MR1, v2015.1
                          B&S Global 544, 555
                          Mitutoyo A504 w/PC-DMIS
                          Since 1994
                          (Advice from my dad.)