B&S Global Advantage 09-15-08, new in oct 2005
0,003 Volumetric (720 mm lenght, max deviation in calibration report)
Volumetric accuracy--where are you at?
Collapse
X
-
B&S Global Status is at .0089 mm on 9/05, spec on machine is .0112 mm.
Leave a comment:
-
This is my calibration what do you think? is this good or bad
RUN X POSITION Y POSITION Z POSITION
----------------------------------------------------------
1 -0.0003 0.0016 -0.0006
2 -0.0006 0.0013 -0.001
3 -0.0004 0.0014 -0.0006
4 -0.0004 0.0015 -0.0008
5 -0.0002 0.0011 -0.0008
6 -0.0003 0.0015 -0.0008
7 -0.0002 0.001 -0.0008
8 0 0.0009 -0.0004
9 -0.0003 0.0015 -0.0009
10 -0.0006 0.0015 -0.0007
RANGE 0.0005 0.0007 0.0006
B89 Sphere Repeat Specification = 0.0015 mm
B89 Repeat Test Has PASSED
Leave a comment:
-
Cal done 9/1/05 .0046 volumetric
B&S 7-10-7
Clean area temp controlled to within 1/2 degree F
Leave a comment:
-
We're good...
Cal done a couple of weeks ago and I'm within my 18 micron envelope...
B&S Xcel 15-20-10
Leave a comment:
-
I'm in! Cal done yesterday.
B & S Xcel 9-15-9 uha
spec. 7.7 cal at 6.8
Leave a comment:
-
Volumetric accuracy--where are you at?
We recently had our calibration done, and due {I think} to several years of use and the installation of a hoist over the machine last year, we had a bunch of gunk in the bearings.
Our tech did clean some of them, and got us down to a 12.2 micron volumetric calibration (outside of the 8.7 manufacturers spec for our machine). Based on what we do, and tolerances etc., we accepted this value for now.
(We will do a thorough cleaning and re-laser, re-cert later this year).
But my question, is how many of you out there are in the same position, outside of the manufacturers spec for volumetric accuracy of the machine?
Just curious........Tags: None
Related Topics
Collapse
-
by MIKEYAny1 knows a good and affordable cal. service located near Houston TX.
-
Channel: PC-DMIS for CMMs
04-23-2008, 05:18 PM -
-
by Wes CiscoI am looking into Pc-Dmis NC. I have read every thread in this sub forum. I have two main question areas left.
Which company(s)...-
Channel: PC-DMIS NC
11-02-2012, 09:34 AM -
-
by TestedGot called over to one of my other plants for a problem with things checking way out of tolerance.
I got there and we watched as the program...-
Channel: PC-DMIS for CMMs
03-01-2017, 11:55 AM -
-
by Paul SarrachAfter I put the call into B&S and called me back the next day, they worked on it for 2 days. And this is thier reply by email no less:
...-
Channel: PC-DMIS for CMMs
10-11-2007, 11:37 AM -
-
by Braz60New to the Forum we were thinking of purchasing an Artifact to verify that our CMM's are still within calibration. Has anyone had experience using one...
-
Channel: PC-DMIS for CMMs
11-22-2010, 03:33 PM -
Leave a comment: