    Default Validation (or calibration) of SmartScope

    Hello,

    We currently have an OGP SmartScope vision system. It is calibrated on a yearly basis by a guy that comes in. I was wondering how this vision system can be quickly calibrated and checked for validation. How do you guys validate that the readings the SmartScope is giving you are correct?

  2. #2
    Moderator
    • Additional Information
      • Years CMM Experience:
      • 1000
      • Software Version:
      • All
      • Machine Type:
      • All
      • Location:
      • Hexagon, Dallas, Texas
      • Job Function:
      • Applications Engineer
    Josh Carpenter's Avatar
    Join Date
    Jun 2011
    Location
    Hexagon, Dallas, Texas
    Posts
    771

    Default

    Moved to correct sub-forum
    "You never change things by fighting the existing reality. To change something, build a new model that makes the existing model obsolete." - B. Fuller

  3. #3
    Member
    • Additional Information
      • Years CMM Experience:
      • 3.5 PC DMIS - 4 Numerex
      • Software Version:
      • 4.1 / 2009 MR1 for Vision
      • Machine Type:
      • 5,7,5 Global Advantage / 3 Quest 250, Smartscopes
      • Location:
      • Andover Massachusetts
      • Job Function:
      • Sr.QE Technician

    Join Date
    Feb 2008
    Location
    Andover Massachusetts
    Posts
    34

    Default

    This is a 2 part question: First in order to “calibrate” you must have the proper glass scales and standards.
    Validation of measurements is approached using the tried and true Gage Reproducibility and Repeatability process.
    If you are concerned more about how close to your tolerance it measures, than you have to understand the limits of your machine and apply those limits to your expectations.
    We tend to use a method of 3 pieces measured 5 times each , in and out of the fixture, (insure rotation is limited) by 2 separate operators.
    We know the limits of the machine and we build guard bands around the tolerance zones depending on the results of the R&R. Example of guard bands applied to tolerances:

    GRR(%) Guardband (%) on both sides
    ≤20 -
    >20 to ≤30 5
    >30 to ≤40 10
    >40 to ≤50 15
    >50 to ≤60 20
    >60 to ≤70 25
    >70 to ≤80 30
    >80 to ≤90 35

    We measure in MM.

Similar Threads

  1. Software Validation
    By Big John in forum PC-DMIS for CMMs
    Replies: 1
    Last Post: 03-18-2008, 11:24 AM
  2. ogp 500 smartscope to pc-dmis
    By scramer56 in forum PC-DMIS Vision
    Replies: 1
    Last Post: 02-06-2008, 07:16 AM
  3. CCM Program Validation???
    By Gary Fitzgerald in forum PC-DMIS for CMMs
    Replies: 18
    Last Post: 01-13-2007, 11:23 AM
  4. CCM Program Validation???
    By Gary Fitzgerald in forum PC-DMIS for CMMs
    Replies: 0
    Last Post: 01-10-2007, 06:26 PM
  5. PC-DMIS Vision V4.1 adds support for OGP Smartscope
    By nkayWAI in forum PC-DMIS Vision
    Replies: 0
    Last Post: 03-30-2006, 04:29 AM

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •